Hi!
That's my first digital scope: A MSOX3024T + brand new.
Sure - my finding is not a big deal - just kinda blemish ... but perhaps one experienced user knows a way to correct that ... (maybe using remote programming or somewhat)
Even when the scope runs on operating temp, that 2.5V / 1KHz square test signal on Probe Comp terminal has a negative DC offset of 100mV ... 120mV (verified using a second scope). Seems there is no menu option for correction.
Little down side: The scope provides training signals, coming out on the same terminal (and on a 2nd one) and all of them have that negativ offset too ...
Btw:
I'm wondering, that these generation of scopes has no option for "zeroing" any input offset drifting caused by operating/environment temp fluctuations (good old analog scopes are equipped with a pot for DC "Balancing", and some of you may remind yourself on Tek probes, having a GND-Ref push button to check that from time to time).
Anyway ... at most sensitive setting = 1mV/div and warm scope (>30 min) the 4 traces just off by <+/-100uV/CH (still on factory cal); guess, that's ok.