Products > Test Equipment

Keysight's new 34465A (6.5 digit) and 34470A (7.5 digit) bench multimeters

<< < (39/145) > >>

LabSpokane:
HAgilSight isn't perfect to be sure (I have worked with them as a client years ago and have seen good and some bad), but they generally do high quality work. I'm assuming that instead of perfection, that the designers reached the best engineering compromise.  Just like we all do.  :-)

HighVoltage:
Today my Keysight 34470A arrived from a German distributor.
Interestingly it also had this label on the outside of the box, to sell it before May 1st 2015
I have no idea why!

After unpacking, I had it warmed up to lab temperature for about 2 hours and then turned it on.
After a few hours of warmup, I hooked the instrument up to a 2.500 V DC reference.

In comparison:
Both 34401A instruments in the picture are calibrated late 2014
The 34461A was bought calibrated new in December 2013
The ACAL was pressed on the 34470A before measurements got started

Having the extra resolution on the trend chart is really something.

Next, I will hook up my SVR-T 10V reference and keep it running for 24 hours

JohnnyBerg:
@HighVoltage: I like it, especially the results you are showing from the 2.5000V reference  ;D ;D ;D

What was the temperature in the lab? (I suspect 21° ~ 22°)

HighVoltage:

--- Quote from: JohnnyBerg on March 16, 2015, 08:33:33 pm ---@HighVoltage: I like it, especially the results you are showing from the 2.5000V reference  ;D ;D ;D

What was the temperature in the lab? (I suspect 21° ~ 22°)

--- End quote ---
@JohnnyBerg, You are fast!
Yes, Lab Temperature is shown at 21.8 degree C right now.

Dr. Frank:
Congratulations! Really looks nice! Especially compared to these old dogs..
Frank