Author Topic: Keysight's new 34465A (6.5 digit) and 34470A (7.5 digit) bench multimeters  (Read 261067 times)
Offline plesa
Re: Keysight's new 34465A (6.5 digit) and 34470A (7.5 digit) bench multimeters
« Reply #325 on: January 12, 2016, 10:08:22 pm »
I have LabView program for 50k/s digitizing with 34411A, so let me know if it can help.
« Last Edit: January 12, 2016, 10:10:09 pm by plesa »
Offline Sparky
« Reply #326 on: January 12, 2016, 10:19:23 pm »
@plesa: It would be great to try it with 34465A/34470A, and see if allows 50kHz digitizing also.  That way we would have a better idea of possible limitations in BenchVue and not with the instrument.
Does your test program require full LabView installation, or can it run "stand alone" as an application?  I'm happy to test it :)
Sparky
Offline plesa
« Reply #327 on: January 14, 2016, 09:50:44 pm »
I build exe, so it is standalone app.
It creates live view, log data to HDD and create live histogram.
For fast readout do not forget uncheck autozero.
http://www.filedropper.com/agilentdmmloggingutility
Offline Sparky
« Reply #328 on: January 15, 2016, 10:31:12 pm »
@plesa: I tried your program but I get the error shown in attachment when I run it.  Do you know what it is?  Anything I can provide to help identify the problem?
(There was a second error about the directory C:\temp but that was easily resolved by creating the required directory).
Offline plesa
« Reply #329 on: January 16, 2016, 10:13:29 am »
Error should be fixed.
Try
http://www.filedropper.com/loggerwithhistogram
Offline ansgar96
« Reply #330 on: January 17, 2016, 05:47:09 pm »
Hey,
I get this problem when my 34465A is connected to my home-network via LAN. Does anyone else have also problems like this or does someone know whats the problem? LAN settings are the default settings.
Offline HighVoltage
« Reply #331 on: January 17, 2016, 05:51:56 pm »
There used to be a problem with agilentlxiwebservice.exe in an older FW
You can find it in a release notes:
#4650  When mDNS is disabled with no LAN connection, AgilentLxiWebService.exe dies.
But I never had this problem.
Offline ansgar96
« Reply #332 on: January 17, 2016, 07:05:04 pm »
Hey,
Forgot to mention, I have the new 2.11 Firmware installed :/
Best regards
Offline Sparky
« Reply #333 on: January 17, 2016, 07:22:27 pm »
Try <snip>
I cannot download from the new link.  Apparently the download does work after getting through the capcha...  I didn't get a chance to re-test as moving my lab equipment...
« Last Edit: January 30, 2016, 10:25:30 pm by Sparky »
Offline bson
« Reply #334 on: February 12, 2016, 08:26:50 am »
Is this a know bug?
Minimum is 10.00000 yet the "Autoscale Once" function places the entire trend graph below 10...
The statistical average positively doesn't match the plot.
Plot is set to show all points.
34465A, firmware is 02.08 and I'm wondering if updating is known to fix this...
« Last Edit: February 12, 2016, 08:29:41 am by bson »
Offline LA7SJA
  • Country: no
  • Acting user manual reader & forum search engine
« Reply #335 on: February 12, 2016, 11:43:08 am »
I can't replicate this on my 34461A running 2.11 but here is the changelog for 2.11 (#2 could be of interest)
34460A/34461A/34465A/34470A Revision 2.11 Enhancements and Fixes
1. Resolved an issue which caused capacitance measurements to intermittently fail with high offset voltages present.
2. Removed/resolved a 3 ppm reading shift that occurred in DCV measurements on the 100V and 1000V ranges when Input Z (impedance) is switched between 10M and Auto.
3. Resolved a Windows Application Error that occurred when LAN service VXI-11 was disabled and power is cycled.
4. Resolved a lock-up condition that occurred during frequency measurements using trigger settings TRIGger:DELay MIN and *TRG.
Firmware, release notes and tools:
http://www.keysight.com/main/software.jspx?ckey=2367633&lc=eng&cc=TH&nid=-536902435.1058513&id=2367633
Johan-Fredrik
« Last Edit: February 12, 2016, 03:22:20 pm by LA7SJA »
"If at first you don't succeed, skydiving is probably not for you"
Offline Dr. Frank
« Reply #336 on: February 12, 2016, 12:01:59 pm »
This bug should have been fixed with 2.08 already:
34460A/34461A Revision 2.08 Enhancements and Fixes
4. Resolved a trend chart scaling defect in which readings would incorrectly appear outside of the vertical scale limits.
The 2.11 fix #2 is not related to the trend chart display, and only shows up in the 100V and 1kV DC ranges.
It's a real measurement fault, which can be displayed in the trend chart as a 3ppm step up/down, each time one toggles the Impedance switch.
Therefore I recommend to upgrade to 2.11, also to make sure, that any known scaling errors are removed.
Frank
PS: Another ~ 10ppm error in the 100V and 1kV DCV range has been found, due to erratic setting of the divider heater.
This has been confirmed by Keysight and is currently under investigation for FW correction.
I hope, that this is also the root cause for the ~ +10ppm ACAL error in these ranges.
« Last Edit: February 12, 2016, 12:12:24 pm by Dr. Frank »
Offline HighVoltage
« Reply #337 on: February 12, 2016, 03:50:06 pm »
I can also confirm, that this is not a problem anymore in the latest firmware.
This was really troubling me in my first 34461A and I am glad it was solved.
Edit:
Added picture of a trendchart from a 10V reference
« Last Edit: February 12, 2016, 03:55:25 pm by HighVoltage »
Offline bson
« Reply #338 on: February 13, 2016, 07:49:27 am »
02.11 looks much better!  But maybe still not quite perfect...
The average is 10.000022 but assuming the Y axis labels are 10.00000, 10.00010, 10.00019 then the graph is too high.  It shows points at 10.000040 and higher, which is above the max value recorded.  This is the case even if the bottom tick is actually at 9.99999 - but then we have TWO errors instead of one...
« Last Edit: February 13, 2016, 07:52:25 am by bson »
Offline Dr. Frank
« Reply #339 on: February 13, 2016, 09:55:48 am »
Well, the problem is, that the graphs scale has 5 digits of resolution only.
So your assumption is very probably wrong. The scaling is 10.00000, 10.00005, 10.00010.
10.00019 as an upper limit would have been rounded to 10.0002, instead.
I recommend to manually scale the y-axis, so you will have better control of the limits.
Frank
Btw.: Received FW 2.12 today. Problem fixed, but not yet the ACAL issue.
Offline Dr. Frank
Re: Keysight's new 34465A (6.5 digit# and 34470A #7.5 digit) bench multimeters
« Reply #340 on: February 13, 2016, 11:22:36 am »
If you open the Vertical Scale menu, then Low = 10.000 0 V, High = 10.000 1 V, that's a 100µV span only. See first two screenshots.
It's not possible to manually choose a higher resolution, because number entry is also limited to 6 digits, for 10V, that's 5 decimal places only.
So there's no error with your instrument, obviously.
At 9.999 98 V input voltage, you will get higher resolution, and might specify Center = 9.999 98V, Span = 10.0000µV
PS: Or even more resolution, i.e. 1 tick = 250nV (Center = 9.999 99V), recorded at NPLC 100.
You can even distinguish the A/D's resolution of 100nV (1E-8).
It's also a '465A.
Frank
« Last Edit: February 13, 2016, 11:41:40 am by Dr. Frank »
Offline TheSteve
  • Living the Dream
« Reply #341 on: February 13, 2016, 06:12:37 pm »
It's nice to see they can be very reactive to the bugs that you find. Please do the 34461A owners a favor and ask them to fix the continuity bug.
« Last Edit: February 13, 2016, 06:15:50 pm by TheSteve »
VE7FM
Offline bson
« Reply #342 on: February 13, 2016, 08:12:07 pm »
Ah, yes, 10.00005 could have been rounded to 10.0001 and this in fact very likely the case.
It would be nice if they'd add another digit, which should fit easily.
Offline sotos
  • Country: gr
« Reply #343 on: February 13, 2016, 11:03:55 pm »
Offline rosbuitre
  • Country: ar
« Reply #344 on: February 14, 2016, 11:11:08 am »
We do not appear in the web   :-//
My instruments: DMM Keysight 34461A / Tektronix DMM916 / Fluke 12, Rigol DS1074Z, Deer DE-5000, Siglent SDG805 / SDP3303D, Dayton Dats2
Offline sotos
  • Country: gr
« Reply #345 on: February 14, 2016, 11:30:25 am »
« Last Edit: February 14, 2016, 11:37:13 am by sotos »
Offline rosbuitre
  • Country: ar
« Reply #346 on: February 14, 2016, 12:34:20 pm »
Here we are.
http://www.keysight.com/owc_discussions/thread.jspa?threadID=40301&tstart=0
And here
https://www.eevblog.com/forum/testgear/34461a-problem-after-updating-to-firmware-a_02_09-behavior-in-continuity-test/
I did not say well  |O , does not appear to download in Keysight web the new version of firmware (2.12)
Regards
Offline Dr. Frank
« Reply #347 on: February 17, 2016, 10:21:19 pm »
Well, your wish is my command..
Jim Durr just confirmed, that they've already solved this known problem, and it will be on the next FW release.
Watch for the Keysight Technology Support site, he promised to post that information there, too.
Frank
« Last Edit: February 17, 2016, 10:24:30 pm by Dr. Frank »
Offline TheSteve
  • Living the Dream
« Reply #348 on: February 17, 2016, 10:33:24 pm »
Frank
Dr Frank to the rescue, thank you sir.
VE7FM
Offline Dr. Frank
« Reply #349 on: February 17, 2016, 10:39:11 pm »
Please, that's not my merit at all.. I just asked Jim Durr about this issue, and he told me that they already had solved it..
Anyhow, it will take some time for the official release, due to their strict Q.A. system, and PV process, i think