If you open the Vertical Scale menu, then Low = 10.000 0 V, High = 10.000 1 V, that's a 100µV span only. See first two screenshots.
It's not possible to manually choose a higher resolution, because number entry is also limited to 6 digits, for 10V, that's 5 decimal places only.
So there's no error with your instrument, obviously.
At 9.999 98 V input voltage, you will get higher resolution, and might specify Center = 9.999 98V, Span = 10.0000µV
PS: Or even more resolution, i.e. 1 tick = 250nV (Center = 9.999 99V), recorded at NPLC 100.
You can even distinguish the A/D's resolution of 100nV (1E-8).
It's also a '465A.
Frank
TheSteve:
--- Quote from: Dr. Frank on February 13, 2016, 09:55:48 am ---Btw.: Received FW 2.12 today. Problem fixed, but not yet the ACAL issue.
--- End quote ---
It's nice to see they can be very reactive to the bugs that you find. Please do the 34461A owners a favor and ask them to fix the continuity bug.
bson:
--- Quote from: Dr. Frank on February 13, 2016, 09:55:48 am ---The scaling is 10.00000, 10.00005, 10.00010.
--- End quote ---
Ah, yes, 10.00005 could have been rounded to 10.0001 and this in fact very likely the case.
It would be nice if they'd add another digit, which should fit easily.
sotos:
--- Quote from: TheSteve on February 13, 2016, 06:12:37 pm ---
--- Quote from: Dr. Frank on February 13, 2016, 09:55:48 am ---Btw.: Received FW 2.12 today. Problem fixed, but not yet the ACAL issue.
--- End quote ---
It's nice to see they can be very reactive to the bugs that you find. Please do the 34461A owners a favor and ask them to fix the continuity bug.
--- End quote ---
:-+
rosbuitre:
--- Quote from: sotos on February 13, 2016, 11:03:55 pm ---
--- Quote from: TheSteve on February 13, 2016, 06:12:37 pm ---
--- Quote from: Dr. Frank on February 13, 2016, 09:55:48 am ---Btw.: Received FW 2.12 today. Problem fixed, but not yet the ACAL issue.
--- End quote ---
It's nice to see they can be very reactive to the bugs that you find. Please do the 34461A owners a favor and ask them to fix the continuity bug.
--- End quote ---
:-+
--- End quote ---
We do not appear in the web :-//
Navigation
[0] Message Index
[#] Next page
[*] Previous page
Go to full version