Hi,
I need some help to understand some datasheet parametes.
Looking at the datasheet for Rohde & Schwarz NGM202, the readback accuracy for current is:
±(% of output + offset): 10 mA range: < 0.05 % + 15 μA
The resolution in the same range is 10 nA.
I interpret this as if my load is known to be 15 uA, the readback could actually display 0uA.
What is the point of having 10 nA resolution when the offset is many orders of magnitude higher?
Using the display to read a sleep mode MCU current would also not make much sense, given that interpretation I have.
Any clarification is much appreciated, thanks.