Large Offset on Rigol DS1054Z - all channels
skua47:
Hi,
I am seeing large offsets in the data collected on my DS1054Z (SN: DS1ZA163554597).
The table below shows the offsets in the data. I used the average math function on the scope to calculate the offset.
Note that the POS dial is set to 0, except for the smallest range. Changing the POS value makes no difference in the offset and turning the AC coupling does not affect the observed offset. I checked with another scope and with a multimeter, the function generator does not account for the offset.
The attached graph shows the result when the same input is presented to both CH1 and CH2. As you can see, both channels have the offset, but the offset on Channel 1 is most severe.
Is there a procedure to calibrate the scope inputs or does the scope require repair? What can you suggest to correct this issue.
Offset (mV)
Range (V/div) Channel 1 Channel 2 Channel 3 Channel 4
10 875 630 1001 780
5 980 379 654 540
2 940 379 399 160
1 1010 391 367 184
0.5 1040 370 390 160
0.2 17 82 13 19
0.1 19.3 16 8 .6 11
0.05 17.7 9 6 5
0.02 20 7.7 7.3 4
0.01 20.4 8 7.8 4
0.005 20.8 7.7 7.7 3.3
0.002 20.8 7.6 7.7 3.1
0.001 20.5 7.6 7.8 3.1
Small shift in offset with BW limit on or off - These are for BW limit off
1X probe, normal mode
Thanks.
David
ataradov:
There is internal calibration procedure that you can run from the menus. If that does no help, then there is not much you can do.
Fungus:
--- Quote from: skua47 on December 29, 2022, 02:27:22 am ---Is there a procedure to calibrate the scope inputs
--- End quote ---
Yes.
Read the manual.
skua47:
Thank
You
Very
Much!
skua47:
Thanks,
I finally found the procedure listed in the manual. The calibration seemed to work. I am really surprised though and how far out the offset was. 1 V seems entirely unacceptable for a commercial instrument.