| Products > Test Equipment |
| LCR ET4410 - Calibration doesn't seem to work |
| (1/1) |
| MichaelB:
I hope I'm posting this in the right spot - I'm not too bright on these blogs... I purchased a new ET4410 from Banggood about 2 weeks ago. It seems to work fine (i.e. makes reasonable measurements) - but... I need help with the calibration. I'm following the manual's (4.7 Calibration) directions for a short calibration. I connect the Kelvin leads and Pres "CAL" to enter calibration - but, absolutely nothing happens. I don't get the "SHORT" picture that's in the manual and there is no countdown. I tried rebooting the machine - rinse/repeat - but still no calibration. Am I missing something? I thought it might be obvious and someone might have seen this before. Thanks! |
| Martin72:
Hi, With ET44/4510 it is as follows: If you want to do a short calibration, short-circuit the test lead and press the CAL button - and hold it down until the short cal message appears. A timer then runs backwards from 20 to zero. -If you want to do an Open Calibration, leave the test leads open and press the CAL button - and hold it down until the Open Cal message appears. |
| MichaelB:
YEP - you nailed it..." and hold it down" was NOT in the instructions, and that was the most important part! LOL. I had to hold it for about 3-5 seconds before it finally entered the CAL mode. I hope this helps someone else. Thanks! |