The price of the DE-5000 is very good for a unit with its performance.
This is due to the use of a dedicated chipset, the Cyrustek ES51919, which I believe emulates the measurement of the old G-R "Digibridge".
My only regret is that it will not display magnitude and phase of impedance, which I believe is a matter of firmware.
One can build ones own test fixture, with true 4-terminal connections, by starting with a TL-21 probe and re-wiring it for different connectors.
Notes:
1. The guts are "floating" with respect to a common terminal, brought out as guard. Since the unit is battery-powered, placing it on a plastic box allows measurement of circuits that are grounded (watch out for charged capacitors or powered circuits).
2. The "calibration" of this and similar units is not calibration of the ohms or farads, since that is due to an internal precision resistor, but a computational correction for the test fixture in use. Even using the panel connectors directly, one should always calibrate the unit after power on with open and short circuits.
3. Since I have to calibrate it anyway, I never minded having the unit not remember the previous setting.
4. The test voltage amplitude is fixed (no user control) at roughly 600 mV rms.