Just for fun we took a 2mm dia 5mm long Copper wire for measurement. This calculates as 1.74nH (Grover) and 0.0267 milli-Ω.
First was using the TH2830 LCR Meter and TH26009B Tweezers @ 100KHz, we couldn't get any type of stable readings and the results wandered from 10~13nH and 0.05~0.5Ω based upon hand pressure, orientation, and repeated DUT engagements.
Then we used the SMD Fixture as shown with the TH2830 LCR Meter, here the results were much more stable and much closer to the computed values but still some slight uncertainty. Note the test current and DUT voltage.
If the DUT were flat we could use our custom Split-Kelvin Fixture for even better results, which was specifically developed for these low Z measurements.
We just can't see how any type of tweezers, even dedicated LCR Meter quality types like the TH26009B can achieve accurate repeatable results at impedance extremes. We've tried over the years to achieve this without success and why we developed the custom LCR Meter Fixtures and modified SMD Fixtures to improve measurement results.
Anyway, guess our skills, techniques and equipment aren't up to par with those articles and those folks here, that can achieve these remarkable results with Tweezers at impedance extremes

Best