Greetings,
I have a Siglent SDG1032X Signal Generator and a Siglent SDS1202X-E Oscilloscope which are both less than 6 months old.
I just noticed something interesting today...
When I have the signal generator produce a waveform with equal positive and negative amplitude, such as a sine or square wave, the oscilloscope will measure that waveform with sightly higher positive than negative amplitude. For example, let's say I set up a simple sine wave having +500 mV and -500mV amplitude. The oscilloscope will show this as roughly +500 mV and -480 mV (not actual measurements... just an example).
I've verified no signal offset or phase angle is set in the signal generator waveform.
I've run the self test on the oscilloscope with no change.
It doesn't seem to matter whether I use an 1x, 10x probe, or a 36" length of RG-58 coax (properly terminated of course) between the signal generator and oscilloscope.
Frequency and signal amplitude don't seem to matter, although the percentage of error changes with both.
Does this sound like normal behavior? Is there anything else I should check?
Thank you,
John
AA7US