Like Bram said above, modern AWGs are quite flat in low frequency range.
Also people questioned use of (well done) RMS measurement with modern 12 bit (or more) scope as a replacement for old thermal RMS voltmeters that are harder and harder to get in good order, if you already don't have one from long ago.
Used was SDG6052X as signalgen. Scope was 1GHz SDS6000H12, roughly equivalent to SDS3000xHD for this test.
180mV RMS sine, swept from 100 Hz to 5 MHZ in step mode (new function on SDG6000X in last FW). 50 Steps in 113sec.
On scope P-P and AC RMS (Stdev) measurement was used together with FFT in peak mode. Flattop window.
Combined flatness for AWG/scope combined is better than 16mdB (0,016dB) from 100Hz to 5 MHz.
I also tried Picoscope 4262, that is 5 MHz 16 bit scope.
With same AWG.
To 1MHz it was 0.1dB, to 2 MHz 0,2dB. -1dB point was at 3.5MHz.
I would say those are good enough results for many uses.