| Like oscilloscopes, can other pieces of bench test equipment create ground loops |
| (1/1) |
| ballsystemlord:
Like oscilloscopes, can other pieces of bench test equipment create ground loops? Like bench-top DMMs, LCR meters, DC power supplies, and load banks? Thanks! |
| tautech:
Bench DMM's, no. Otherwise pretty much any instrument with a grounded BNC connector one must be wary how they use it. Some DC supply's need be watched too as outputs are not always isolated. |
| David Hess:
Power supplies usually have floating outputs, but not always. Most other instruments with BNC connectors will create ground loops, but there are some exceptions. For instance some function generators have a galvanically isolated output. |
| jonpaul:
Wrong assumptions, so title makes little sense. Check your def of "ground loops" j |
| David Hess:
Below shows an example of a pulse generator which makes some attempt to minimize ground loops. The shell of the output connector is isolated from the chassis. |