Products > Test Equipment
Like oscilloscopes, can other pieces of bench test equipment create ground loops
(1/1)
ballsystemlord:
Like oscilloscopes, can other pieces of bench test equipment create ground loops?
Like bench-top DMMs, LCR meters, DC power supplies, and load banks?

Thanks!
tautech:
Bench DMM's, no.
Otherwise pretty much any instrument with a grounded BNC connector one must be wary how they use it.
Some DC supply's need be watched too as outputs are not always isolated.
David Hess:
Power supplies usually have floating outputs, but not always.

Most other instruments with BNC connectors will create ground loops, but there are some exceptions.  For instance some function generators have a galvanically isolated output.
jonpaul:
Wrong assumptions, so  title makes little sense.

Check your def of "ground loops"

j

David Hess:
Below shows an example of a pulse generator which makes some attempt to minimize ground loops.  The shell of the output connector is isolated from the chassis.