Wouldn't a TRMS AC+DC meter essentially be flat to DC?? Hold on, I'll try one.
EDIT: TRMS AC+DC doesn't work so well below 10 Hz because I can't slow down the sample rate and thus the reading fluctuates, as you'd expect. So, a quick low frequency test of a few meters set to AC using a 2.000V signal:
Fluke 8846A w/ filter set to 3 Hz reads pretty much spot on down to 3.0 Hz, and then drops off starting at 2.9Hz. I guess when they say 3Hz they really mean it. This is supposedly a 3 or 4 pole filter, I don't remember which. Accuracy is specified to 3 Hz. 3.0000 Hz, that is.
Fluke 289 stays accurate down to 10 Hz and then fluctuates due to sample rate. Accuracy specs go to 20Hz.
Fluke 189 stays accurate to 5 Hz, then fluctuates. It isn't 'better' than the 289, but has a slower sample rate. Accuracy specs go to 45Hz.
Simpson 270, reads nicely and probably more than good enough for subwoofer tuning--the error is a LOT less than 1dB, probably more like 0.1dB. Below 10Hz the needle starts to shiver, but it is accurate and reasonably readable down to 2 Hz. Suprisingly, the Simpson is specified for AC from 20Hz to
100kHz. Apparently the needle shivering is the only thing limiting the lower end.
So the winner is the Simpson.