Jumps J8 and J10 are a stable 10.004 kOhms resistance.
J7 and J12 can keep a relatively stable capacitance of 156uF.
(other combinations also work with different values) (turn it off first)
J-R:
If you want a budget reference you can rely on I would suggest the excellent VREF-01 from voltagestandard.com for only $31. Calibrations are free for the first 2 years.
faraday:
I tested this reference over 24h with 34410A and 33401A and so far is very stable.
12H 24H
34410A 9,85795 9,85795
34401a 9,85787 9,85784
Trader:
--- Quote from: faraday on February 04, 2021, 08:01:20 am ---I tested this reference over 24h with 34410A and 33401A and so far is very stable.
12H 24H
34410A 9,85795 9,85795
34401a 9,85787 9,85784
--- End quote ---
Awesome. Test the Current too:
https://www.eevblog.com/forum/testgear/lm399-voltage-reference-fluke-289-gossen-29s-rm303-an870/msg3394242/#msg3394242
faraday:
Voltage / time use BNC to banana cable
Navigation