What is the maximum signal frequency that a Logic Analyzer / MSO Digital input can capture based on Nyquist criteria?

I tested with a 500MSa/s Agilent Logic Analyzer and an MSO with the same 500MSa/s on digital inputs.  I generated a 112MHz clock signal using the Si5351 clock generator and the MSO analog input measures correctly 112MHz, while the digital input measures 100MHz.  The Agilent Logic Analyzer shows the same waveform distortion (not equally spaced high/low square wave).

Is it reasonable to say that it requires 5 times the signal frequency to sample the correctly even when the waveform is distorted?  In this case, 500MSa/s / 5 = 100MHz.

MSO: 112MHz
[attachimg=3]

Agilent 1671G: 112MHz
[attachimg=1]


MSO: 13.51MHz
[attachimg=2]
 
The following users thanked this post: Electro Fan

Online Tomorokoshi

  • Super Contributor
  • ***
  • Posts: 1027
  • Country: us
Disclaimer: I haven't used an HP Logic Analyzer for a long time.

1. Are you in State or Timing acquisition mode? There are limitations around 150 MHz for State mode. Timing mode has limitations of 250 MHz or 500 MHz depending on the channel configuration of the pods.

2. With TTL thresholds the sine wave might not be achieving setup and hold requirements, so it could be missing cycles. Is it possible to drive a square wave of your frequencies into it?
 

Offline TK

  • Super Contributor
  • ***
  • Posts: 1671
  • Country: us
  • I am a Systems Analyst who plays with Electronics
Disclaimer: I haven't used an HP Logic Analyzer for a long time.

1. Are you in State or Timing acquisition mode? There are limitations around 150 MHz for State mode. Timing mode has limitations of 250 MHz or 500 MHz depending on the channel configuration of the pods.

2. With TTL thresholds the sine wave might not be achieving setup and hold requirements, so it could be missing cycles. Is it possible to drive a square wave of your frequencies into it?
1. I am using timing acquisition mode @ 500MHz
2. It is a square wave.  Both the logic analyzer and the MSO are using TTL threshold.  With CMOS or LVCMOS the waveform changes, but still not regular high / low transitions. 
 

Online Tomorokoshi

  • Super Contributor
  • ***
  • Posts: 1027
  • Country: us
Ok. On the MSO, the analog waveform seems to be slew-rate limited / filtered to be something between a sine wave (which I had originally assumed) and a triangle wave.

Could there be excess parasitics with the cabling setup? It sounds like from the original posting that both instruments were on the signal generator at the same time. That is, the yellow trace on the MSO may be showing the actual waveform going into the logic analyzers.

1. Can you try one instrument at a time?

2. It looks like the source impedance should be 50 ohms. What are the cable and input impedances of the instruments? How long are the cables, what type, etc.
 

Offline TK

  • Super Contributor
  • ***
  • Posts: 1671
  • Country: us
  • I am a Systems Analyst who plays with Electronics
Analog channel has 200MHz 10X probe.

1. I had 1 instrument at a time (digital channels on the MSO)... I then connected the analog channel to verify.
2. The SI5351 board is connected to a breadboard, but I am using male pin headers to connect the analog probe and the digital POD cables. 

With the same setup, it can show a nice square wave on 45MHz SPI SCK signal.  I need to test increasing the clock to see when I start seeing irregular (not perfect 50% High-Low) signal on the digital inputs.  The MSO has 2GSa/s for 4 channels, I was using 2 channels at the time of the measurement, so it was sampling @ 1GSa/s.  Digital inputs are sampled at 500MSa/s.
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 22029
  • Country: nz
  • Taupaki Technologies Ltd. NZ Siglent Distributor
    • Taupaki Technologies Ltd.
IMO the impedance of the MSO is impacting on the result whereas a 10x probe will have less effect.
While 10x probe vs MSO capacitance loadings are similar the R of the MSO probe is a magnitude less.

I might be wrong but I see the MSO insertion having a greater impact on a result which of course has little to do with Nyquist.
And of course each bus will be impacted differently depending on how low its impedance is.
Avid Rabid Hobbyist
 

Online nctnico

  • Super Contributor
  • ***
  • Posts: 21456
  • Country: nl
    • NCT Developments
What is the maximum signal frequency that a Logic Analyzer / MSO Digital input can capture based on Nyquist criteria?

I tested with a 500MSa/s Agilent Logic Analyzer and an MSO with the same 500MSa/s on digital inputs.  I generated a 112MHz clock signal using the Si5351 clock generator and the MSO analog input measures correctly 112MHz, while the digital input measures 100MHz.  The Agilent Logic Analyzer shows the same waveform distortion (not equally spaced high/low square wave).

Is it reasonable to say that it requires 5 times the signal frequency to sample the correctly even when the waveform is distorted?  In this case, 500MSa/s
You have to look at this from the time domain: The digital sampling interval needs to be smaller than the narrowest pulse you try to capture. Digital probes usually have enough bandwidth; even with attenuated signals there is more than enough swing to meet the minimum hysteresis requirements. Worst case you'd have to adjust the thresholds a little bit.
« Last Edit: May 13, 2020, 06:47:11 am by nctnico »
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 
The following users thanked this post: TK