Author Topic: Looking for a fast booting bench DMM  (Read 5705 times)
Offline HKJ
    • Tests
Re: Looking for a fast booting bench DMM
« Reply #25 on: May 14, 2021, 11:26:34 am »
I agree! If accuracy isn't important then why use a bench DMM? They don't offer any more functionality than a hand held one except accuracy.
I may have gotten the wrong handheld meters and even believed that I had the most advanced on the market, but none of my handheld meters can match my bench meters in functionality.
Offline bateau020Topic starter
« Reply #26 on: May 14, 2021, 11:33:10 am »
.. plus less clutter on the desk. The handhelds tend to get in the way while working on a project. Many desks (mine at least) are always too small.
Online Kleinstein
« Reply #27 on: May 14, 2021, 11:44:51 am »
4 wire ohms is very rare with handheld meters. Another points is input impedance much higher than 10 M. Most of the other functions can be found with handhelds.
Modern DMMs usually do a more or less extensive self test after turn on. This takes it's time as the circuit needs to stabilize first. Especially more modern meters takt the self test more serious, which is normally a good thing. Still faster boot time would be nice and not that difficult. However an OS like Linux with a relatively slow CPU takes it's time.
Ethernet or Wifi are 2 rather modern parts, which excludes old meters.
Offline HKJ
    • Tests
« Reply #28 on: May 14, 2021, 11:50:54 am »
Not really, bench meter can sample much faster, show trend charts (A few handhelds support this), histogram and more statistic than handhelds.
Handhelds have a higher CAT rating than bench meters and some can do stuff that bench meter cannot do.
Offline BeBuLamar
« Reply #29 on: May 14, 2021, 01:08:59 pm »
Both 4 wire resistance measurement and high impedance above 10M are about accuracy. They improve the accuracy of the measurement not additional functions.
« Reply #30 on: May 14, 2021, 07:11:21 pm »
With a heavy heart I come before you, to tell you that I have succumbed.
My search for a good bench multimeter has lead me astray. I became enchanted by that, which I have sworn to avoid. I have just bought myself a slow booting bench multimeter. A keysight 34465A.   :-[
The specs were just too much for me. That machine will finally allow me to more or less properly work on battery operated devices. I have decided to re-evaluate my priorities. Physical size and current sensitivity are finally more important to me than boot time. I'll get over those 25-30 seconds. And the price. With the free memory expansion on top of it, I will have gifted myself with a machine that will last me some years. I just hope that the change of OS that some rumour to be coming (the version of Windows CE that is embedded being EOL probably next year) will not suddenly make this machine obsolete. Now on for the hunt for some good kelvin probes.
By the way, for future reference, here is my shortlist (and yes, machines of various categories in here, subjectively sorted from good to questionable):
(only new, since second hand market in europe is very strained)
Clones are not mentioned. And only meters with 4W and built-in ethernet connectivity are mentioned. A meter that has no ETH is outdated.
* Yokogawa DM7560: fast boot (don't know how much, but fast), excellent specs and features, but hard to get here, and too deep for my bench.
* Tek DMM6500: 15 sec boot, too deep for my bench, otherwise very nice.
* Keysight 34461A: 25-30 sec boot, but the 34465A has higher current sensitivity.
* Fluke 8846A (aka Tek DMM4050, but no longer sold new): 1 sec boot, but old design, repairs seem difficult
* BK 5492C: boot time unknown, 1.2M count, too deep for my bench
* Siglent SDM 3065X: 17 sec boot, 2.2M count, too deep for my bench
* GW-Instek GDM9061: 1 sec boot, 1.2M count, but noisy (electrically) it seems
* Siglent SDM3055X: 17 sec boot, 240k count, otherwise very nice for the price.
* HMC 8012: 11 sec boot, 480k count, lower amp range not low enough for me, secondary measurements seem hampered.
* AIM-TTI 1908P: 7 sec boot, 120k count, too slow sample rate (1/s), lower amp range not low enough for me
* OWON XDM3051: 10 sec boot, 200k count, dual display but no DCV+DCI, looks very price-optimised.
I've got more details about most of these meters of course, like ppm, price, sample rate and power consumption. If you want me to add that, can do, but most of these meters are already mentioned in the specialized multimeter topic (and selected numbers never do justice).
« Last Edit: May 14, 2021, 08:17:24 pm by bateau020 »
The following users thanked this post: Markus2801A, narkeleptk
Offline nctnico
« Reply #31 on: May 14, 2021, 10:00:29 pm »
If taking best advantage of the presumed extra accuracy of a bench multimeter means leaving it turned on, then boot time becomes irrelevant.
Sometimes I just want to take a quick measurement, then boot time becomes relevant :).
Yes, we get that. But since bench DMMs are designed for accuracy,
Sorry, but that is utter nonsense. Since the dawn of the electronics industry you can buy bench DMMs which aren't designed for accuracy. There is more to a bench DMM than accuracy: stackability, better readability of the digits and a mains cord so it doesn't need batteries. I don't get why and how people have gotten the idea that a bench DMM should be all about accuracy.  :palm:
« Last Edit: May 14, 2021, 10:07:01 pm by nctnico »
« Reply #32 on: May 14, 2021, 10:10:43 pm »
Yep it's all about needs.
Small repair shop 1000km from me uses 3x SDM3045X on their repair benches that give sufficient accuracy, enough capability and zero fan noise.
Avid Rabid Hobbyist
Offline tooki
« Reply #33 on: May 14, 2021, 11:19:02 pm »
Sorry, I accidentally said “accuracy” when I meant “precision”.
Beyond that, it’s not “utter nonsense”. Look at the bench DMMs on the market: almost none aren’t high-precision, high-speed meters. You imply that they’re just the same as handheld meters but mains powered and with different displays, which is nonsense. Plus, do you see the word “only” in my statement? Nope. Because I wasn’t enumerating what the virtues of a bench meter are. What I was saying is that because (practically all) bench meters are designed for high precision and high accuracy, they rely on high accuracy circuits that need to warm up. That they have other characteristics is irrelevant to the point I was making. I never said bench meters are “all about accuracy”, that’s what YOU chose to hear based on things I didn’t say.
So you can get off your high horse now.
« Last Edit: May 14, 2021, 11:24:57 pm by tooki »
« Reply #34 on: May 15, 2021, 04:41:46 am »
gentlemen, gentlemen. It's all a matter of perspective and needs. The claim that "(practically all) bench meters are designed for high precision and high accuracy" is true for some brands. There are however many brands that propose meters with 20.000 count or less. Aliexpress has a nice selection of them. These meters mostly do not need a long warm up. And there is market for them.
« Reply #35 on: May 15, 2021, 05:00:40 am »
Then off you go as you know the answer  ;)
For most of us here under a few hundred K (more like 1 million+) and 4 wire minimum spec isn't a Bench DMM it is a 'Bench formfactor DMM'. Your initial 'requirement' of rapid boot time simply doesn't gel with proper bench DMM's. These are used for more accurate and hence STABLE readings that stability comes after 10-30 minutes so boot time just isn't a factor.
Also most non specific 4W capable DMM's as distinct from Milliohm meters SUCK at sub ohm measurements so milliohm resolution is also a bit pointless.
Power consumption really  :o
Nothing wrong with screwing a good handheld to the wall and using that if it gets the job done. The pair of junkers below are generally used for USB Temperature logging and sit to one side of my bench. As needed I also use them for other non critical logging jobs from time to time. They are always there but I still use my 6 1/2 digit meters for measurements as they sit on the right side of the bench.
So save your $ and buy a good Handheld with connectivity.
« Last Edit: May 15, 2021, 05:09:18 am by beanflying »
Offline nctnico
« Reply #36 on: May 15, 2021, 07:55:29 am »
All bench meters? No. Just go around and count the models. You'll see there are way more models with less than 5.5 digits compared to models with 5.5 (or more) digits. Or if you want to go by accuracy / precission (words which mean about the same so let's not start a semantic discussion) better or worst than 0.02% . You just have to look in the right places.
« Reply #37 on: May 15, 2021, 08:54:59 am »
I think I understand the point is this:
What meters are allowed to be called "Bench DMM" and what are not. It is like "Real Men Don't (fill in what you dislike)". Let's just agree definitions/opinions differ. But that is not the point of this thread. This thread did not start out as a "what is a bench meter" question. It started out as "what meters can start quickly". Some can do that, some don't. And yes, although I did not state it, I am aware that for higher accuracy one needs a warm up period. (I don't know if warm-up has an evenly significantly impact on precision though, would be interesting to know.)
To steer this discussion to a more objective direction, are there any meters that use low power standby (<2W should be enough if well isolated) to achieve very short warm-up and startup time? Frankly, that would be nice.
The 8846A, although a fast booter out of standby, still needs a fairly long warm-up AFAIK. I haven't found them. Maybe they don't exist yet.
<rant>
And yes,
Let's not get political here. Power consumption is important to me. Already for the likely related fan noise, as I stated, but also because I try not to be wasteful. Taint me green, yes, I have a solar panel and an electric car. In short, I care. There are more around here that do. My choice. Why would that be bad?
And it IS possible. Some meters pull > 30W, some < 11W. Yes, those numbers exist. Even if it pales to the consumption of a hot air gun, it is still something.
</rant>
« Last Edit: May 15, 2021, 08:58:18 am by bateau020 »
« Reply #38 on: May 15, 2021, 09:08:00 am »
You clearly are missing any points put so if you ask for information don't argue against it take it in. If you claim to KNOW the answers then don't waste our time!
Your opinion and attempted 'required specification' of what makes a Bench meter YOU require is WRONG and shows a lack of basic understanding. Further your requirements read like and are better suited for a good quality handheld and NOT a bench meter.
You need to seriously look at and understand what 4W on a bench DMM is and is not. 4W does not make any meter nesecerilly able to resolve 'accurately' milliohms. My three 6 1/2 digit DMMS are put to shame by my milliohm meter by chunks. 4W is not magic.
Yep I question 'power consumption' used as you suggest quick boot then measure and then presumably turn off how much power are you planning to save?
« Reply #39 on: May 15, 2021, 09:52:54 am »
Torn between not replying to further feed a discussion that is not really helpful and softening the situation a bit.
And I agree if you take into account only a part of all this, and if you take a specific definition of what a bench dmm is.
One of my reasons (that I should have stated more clearly) is that I do not want another handheld; I have enough of them, and some very good ones. I want more desk space. Accuracy is not my prime goal. Hence my quest for a relatively capable and more or less universal tool. Sometimes it will be for 1 single measurement, sometimes it will get prolonged use. Depends on the needs of the moment. Yes, I have a lack of full understanding and no doubt I have false expectations. I do want to learn more about it, it's been a long time since I've used any serious bench meter.
And to make this a bit less personal:
there are plenty bench meters (or bench formfactor meters) for which accuracy is not the ultimate goal. And still, some of them are slow. Hence the thread.
The fact that I chose another meter is for my personal reasons.
« Last Edit: May 15, 2021, 10:25:00 am by bateau020 »
Offline nctnico
« Reply #40 on: May 15, 2021, 11:04:18 am »
I agree here. As an owner of a Keysight 34461A I have to say measuring low resistance is not it's strong suit and it is easely outperformed by other, more specialised instruments.
@bateau020: for your requirements it might be better to buy 2 bench DMMs. A cheap everyday one like the VC8145 (I'm using one as a daily driver for nearly 10 years already) and one with more precission / better interfacing for automated tests. Then you have the best of both worlds. In many cases it is nearly impossible to have one instrument fill all the requirements.
« Last Edit: May 15, 2021, 11:07:20 am by nctnico »
Offline tooki
« Reply #41 on: May 15, 2021, 11:33:14 am »
First of all, STOP PRETENDING I USED THE WORD "ALL". That's blatantly dishonest, as I did not say, and am not saying now, that it's "all".
Beyond that, let's look at a typical test gear vendor, Tequipment. Of 47 bench meters they have, four are 4.5 digit, the rest are 5.5 digit and up. Distrelec: six of 88 are under 5.5 digits. Conrad: 13 of 62. Farnell: 21 of 91.
Reichelt (which is much more consumer-oriented) is the first one I checked that has more <5.5 digit meters: 14 of 24. But being consumer-oriented, they carry much more low-end stuff, and very little high-end, so in isolation, it's not a representative sample.
Sure, if you start including crapola on AliExpress, where low-end stuff dominates and every item is listed 300 times by different vendors under different brands, you'll end up with high numbers, but I don't believe that's even distantly representative of the market, and of market penetration.
Offline BeBuLamar
« Reply #42 on: May 15, 2021, 12:20:29 pm »
How can a bench DMM gives more desk space than a hand held DMM? The OP kept saying he needs more desk space.
« Reply #43 on: May 15, 2021, 12:29:11 pm »
Yes, you're probably right. That or ET3240 or something other cheap. If not only for the startup time of the 34465A probably.  ;)
Got shelves above the desk. Putting the meter there frees up space. A handheld there is not practical since you risk pulling it off the shelf, you cannot see the screen well, and manipulation is cludgy. Unless I screw it to a wall of course.
The following users thanked this post: narkeleptk
Offline nctnico
« Reply #44 on: May 15, 2021, 12:32:48 pm »
Simple. You either have to put a handheld DMM in front of other equipment or lay it flat. Neither is ideal. A bench DMM OTOH can be stacked into a pile of other test equipment with the leads tucked away nicely if you want. The result is a much less cluttered work area. I have gotten rid of all my handheld DMMs except one (which I only use when I need to measure something which is not on my workbench) for this reason.
The following users thanked this post: tooki, narkeleptk
Offline bdunham7
« Reply #45 on: May 15, 2021, 01:59:50 pm »
No, the 8846A does not need any warmup time out of standby--the whole reason it has the standby mode is to keep the LM399 reference heated and stabilized.  And since it is a fanless low-tempco design, it only requires about 15 seconds from a cold boot to be fully ready for use.  I don't know what its standby power consumption is--probably more than 2W--but it doesn't seem to get warm at all in standby, so likely not much.
« Reply #46 on: May 15, 2021, 02:20:35 pm »
ah, the user's manual says "Accuracy specifications are valid for 61⁄2 digit resolution mode after at least a 1-hour warm-up with Auto Zero enabled."
Might be just to cover their behinds.
That meter is interesting indeed.
Offline bdunham7
« Reply #47 on: May 15, 2021, 02:32:13 pm »
4W actually is magic, IMO.  But the issue with very low resistances isn't because the 6.5 digit DMM can't resolve milliohms--they certainly can on a relative basis.  The problem is the zero point, and that is just a result of measuring the very low voltages involved.  So trying to read 1 milliohm with a 10mA test current--the most I've seen from a precision DMM--is the same as trying to measure 10uV and you have all of the same issues.  I can get stable readings down below 1mR, but it takes some careful set up of leads and quite a bit of patience.  A milliohm meter is one way to make this easier, but I just use a bench PSU and move my 4W source leads to the PSU and supply 1 amp--then set the DMM to volts and I have a microohm meter.
If you really want to see magic, look at the Fluke TL2X4W split-jack test lead sets.
Offline bdunham7
« Reply #48 on: May 15, 2021, 07:03:40 pm »
So for all you lurkers, here's a deal.
https://www.ebay.com/itm/154445651615?ul_noapp=true
This is a Fluke 8808A equivalent, so make sure you know what that is before you bid.  5.5 digit 100ppm basic DC accuracy, limited high-impedance range, goes for $800-900 new.  Dual-display with dedicated segments instead of the matrix-type on the 8845A/8846A.
Offline Markus2801A
  • Country: at
  • Pobody’s Nerfect ;-)
    • KEM InfoPage
« Reply #49 on: March 02, 2022, 04:27:47 pm »
GW Instek GDM-9061 (or 9060) boots fast.
I bought one a year ago and haven't used it since (life got in the way) but i was nicely suprised by the quick power on when i tested the unit.
https://www.gwinstek.com/en-global/products/detail/GDM-906x
https://www.eleshop.nl/gw-instek-gdm-9061-multimeter.html
Youtube video:
https://youtu.be/-x8Bvjc8yIQ
Hello Peter,
did you have time to get more in touch with your unit? I really would like to hear your experience with the GW Instek GDM-9061.
Teacher for electrical Engineering @ HTL and Werkmeisterschule :-)