I have a 7A13 diff amp to do the 20MHz BW testing. That one was expensive $50.00it has a interesting infinite impedance mode for 50mV/div and below.
@ muvideo,
Thanks for the heads up on the high impedance mode. I missed that in scanning the manual.
Do your 7A22's have the awesome machined tuning capacitors like mine in this thread?. The manual I downloaded shows regular tuning caps in the manual. The jumpers that get unsoldered for the high impedance mode are shown in those pics also.
That is one kick ass microvolt input scope! Few can match it today, most top scopes do 500uV/div as minimum, if at all. Not to mention the dynamic range, 10uV to 10V /div, it took a while to look at that range before I realized it was true!


IMO don't take it apart. Get the white handle Microbrush to apply the Deoxit. Cotton swabs will waste a huge amount of the very expensive stuff. Just wet each each contact with D100L and then cycle the switch a lot. Wash off with your favorite electronic cleaner and blow out. Then lightly lube each contact with Deoxit G100L with another white microbrush. I competely agree with using the liquid (not spray) versions of deoxit. HERE is my setup.


So here's my progress:
So I'm kind of confused. As you can see, after I cleaned it all the peaks and valleys are gone, which I thought made sense, but then when I calibrated it, they came back, and it's nowhere near as accurate as I thought it would be. Zero didn't really need to be adjusted all that much, it was at about -0.8mV, but I got it to within 0.5uV, then at 20V I got it to 20.005V.
Lets put this in perspective. The spec for the 2020 is 0.1% of setting plus 500µV.
Lets not even consider the additional 0.001% regulation, 8 hour stability of .001% + 100µV, and TC of 0.001% or 50µV whichever is greater/ deg C.
Min: 5.00231 that is 40µV variation (.36% of a .011V error band) Max error is .00235V spec allows .0055V less than 42% of spec.
Max: 5.00235 +.047% of set value error.
Min: 0.500546 that is 46µV variation (2.3% of a .002V error band) Max error is .000592V spec allows .001V less than 59% of spec.
Max: 0.500592 +.118% of set value error.
Min: 0.0499536 that is 41µV variation (3.7% of a .0011V error band) Max error is .0000464V spec allows .00055V less than 9% of spec.
Max: 0.0499577 -.093% of set value error.
Min: 0.00498626 that is 9µV variation (0.89% of a .00101V error band) Max error is .00001374V spec allows .000505V less than 3% of spec.
Max: 0.00499526 -.275% of set value error.
This is all well within the original specifications.Those microvolt variations are milli ohm variations in contact resistance. It is fun to try to improve it further.
The green line is expected, the blue line is erratic and suggestive of dirt; the yellow line now appears similar.
I suspect the calibration pot has oxidized grit; once manipulated the dirt is catching on the pot wiper arms. Try turning the adjust multiple times back and forth to clear out the grit and see what happens. I can't recall if the cal pot is hermetically sealed or not, and amenable to dexoit cleaning.
Lets put this in perspective. The spec for the 2020 is 0.1% of setting plus 500µV.Actually I forgot to point out this is my 2005, my 2020B is already pretty good. I'm hoping it can get better after being cleaned.
So, don't worry about the accuracy, just be happy the repeatability and drift are good (I measured it at 6uV at 12 hours so far)?
I suspect the calibration pot has oxidized grit; once manipulated the dirt is catching on the pot wiper arms. Try turning the adjust multiple times back and forth to clear out the grit and see what happens. I can't recall if the cal pot is hermetically sealed or not, and amenable to dexoit cleaning.
The pots aren't sealed as far as I can tell, but I'm not sure how I'd open them without destroying them. I tried wiping them back and forth several times, but it doesn't seem to have made any difference.