gmac34:
Ok, maybe there is some confusion with adjustment and calibration, I think in Italian "calibrazione" is fiddling with pots and "taratura" is comparing with a known standard. Sorry
Anyway what I was looking is both, if needed.
I received the replacement fine-line, measured to see if it was ok, installed, and the issue was still there, I measured the original, it seems good to, so I put it back and the fault was gone. there might have been some oxidation on the golden fingers. So I cleaned everything and reinstalled the original.

I guess it's fixed now and didn't need a new U200 after all.

I will not touch the DCV calibration, it is quite consistently less than 5-6 ppms away from my other meter.

ACV and resistance might still require some fiddling, but I need some better resistors to do something about that