Good thing I told myself not to mess with that, who knows what it would have been otherwise...
Shot 1 is the Sensepeek probe, 2 is a Testec without touching the high speed compensation. I am pretty sure both were overshooting 1% more Friday. I have a screenshot in the internal memory of the scope, but somehow didn't found a way to copy or move it to a thumb drive, so I can't check nor share. I was measuring with both probes on 2 channels at the same time then, not sure if it made the difference.
Not too bad, so call it ok ? Naaah...
Shot 3 is from my "DG992", with a BNC cable to the scope, 50 Ohms termination set on the scope. The BNC cable was bought somewhere, some time ago, with the main criteria being price, so probably not the best cable possible. On the other hand, it is carrying a only 1MHz square signal, with 8ns rise time so it shouldn't have too much influence on the results I guess. There's some ringing, which seems typical of this generator and a bit of undershoot, which I am not sure if usual or not.
Shot 4 is with a Testec, connected with the supplied BNC adapter. This looks a lot like the user manual example for high speed overshoot. While I'm at it, shot 5 is in 1x gain. As expected, slower, exponential rise.
Shot 6 is the same setup, but with the HF compensations (2 screws for that) adjusted so that the waveform looks as square as possible. Shot 7 is the result on the 1KHz signal. Interestingly there the overshoot is higher.
Shot 8 is the Testec probe adjusted to exactly match CH2 on the generator, fed to the scope CH2 via the BNC cable. With 50 ohms termination, the peak peak value is reduced compared to the 1 Vpp setting. I changed the display zoom of CH2 to compensate. I found the matching impressive. Crossing channels on the generator and then on the scope did not change the matching. Is it a more faithful representation of the signal ? No idea. Can you match the behavior of a BNC cable 50 ohms terminated with this 10x probe ? Yes definitely.
Finally, shot 9 is the result on the 1KHz square. Not much of a difference compared to previous HF compensation fiddling in terms of overshoot, a bit slower rise time.
So what do I think I learned here ?
- The bigger your screen, the higher the resolution of the scope and the lower its noise, the more there is to see, both interesting and distracting. Knowledge is power, ignorance is bliss, all that things
- The overshoot on the scope 1KHz terminals seems to be really there, and not just a probing artifact. It's rise time is too slow for the overshoot to be affected by the HF compensation adjustments in a tangible way, and its shape is also unaffected by the usual compensation.
- The Testec probes are nice to use, with good provided tips and grounding options. There is 3 compensation screws which can set to match the response of a direct BNC 50 ohms connection. Not sure if the latter is something I want.
Also, I have 2 more usability remarks on the Magnova :
- As written above, I couldn't find a way to move files around folders, both within a drive and across drives. The manual wasn't helpful either. If someone knows how to do it, I would appreciate.
- When using long time bases (>2s/div) the data isn't displayed while being recorded, but only when the capture is complete. There is a handy count down displayed at the top of the screen showing the pre-buffer filling process before trigger and then the time to finish filling it, but when I'm working on signals <1Hz, I like seeing a real time trace filling the screen.