There is nothing to adjust there. Gigatronics calibrator ref level is set up by setting a CALFAC number. The older meters like 8541\8542 had access to this through menu:
MENU -> SERVICE MENU ->CALIBRATOR -> EEPROM -> SNumb (unit serial number) -> CALFAC change
but their calibrators were fixed 50MHz.
The newer 8651/8651 series apart from 50MHz also have a range of frequencies from 800MHz to 1GHz, option 12. So they may probably have two or more cal factors.
What I need is know is how to access, change and save the CALFAC-s on the newer models. There must be an external programming protocol or hidden service menu or both.
The cal-labs calibrating these meters certainly know how to access the CALFAC-s.