Meter out of range
spacebiscuit:
Sorry if this is a bit of a newbie dumb question, but I have the following multi-meter - only about 1 month old
Just recently I noticed that the display is showing fluctuating result on the dc voltage settings, however it always reset to 0 when I touch the red and black probes which I understand is desireaable. however when I try to measure for example a 1.5v Aa battery the screen display 1 - as if out of range. I've tried the other ranges and get the same. The diode/continuity mode seems to work ok.
Any ideas - I tried the battery in another meter and seems ok.
Fungus:
--- Quote from: spacebiscuit on February 25, 2024, 01:14:09 pm ---Any ideas - I tried the battery in another meter and seems ok.
--- End quote ---
If you've tried all the ranges then it sounds broken to me.
The difference between cheap and expensive meters is how well they survive "events". Did anything bad happen to it?
radiolistener:
Looks like bad contact on mode switch. Most of all it works in 200 mV mode instead of 2 V due to bad contact.
Try to disassemble it and clean contacts with rubber and alcohol.
tunk:
Does the AA cell read correctly in the 20V range?
spacebiscuit:
I'm getting out of range on 20v range too.