Micsig CP1003/CP503 100MHz/50MHz Current Probe
amar87:
There is a Whatspp number on the Micsig website. i wrote to them they also said it can be the slider. I gave them the forum link and invited them to comment as well, maybe they will do.
Martin72:
Your pictures are always opened separately, which is a bit tiring when the post is gone and only the picture is visible.
I think the small increase that can be seen in the "cable test" is negligible, so the graph from the currentprobe is "real".
Micsig should really take a look at this, the currentprobe is too expensive for that.
amar87:
Fo the sake of completeness i also did an 'ugly' bode plot manually:
Generator set to 50R output impedance, probe sldier closed.
The scope is set to 1 Meg input for CH1 where the output of the probe is, 50R CH2 where the generator goes.
Gives this :
Welll..... i gues if you would imagine that that part i circled in blue would be flat like the yellow line, it can be like in the datasheet...
Yes would really be nice if micsig would say something about this.....
amar87:
how to place the pictures so they open inline (sorry for the forum-noobness)?
Martin72:
Hi,
I have recreated your test with the cable only, it's just like that, no need to worry.
It depends on the cable and the internal generator does not necessarily have to be ultralinear.