Products > Test Equipment
Micsig CP1003/CP503 100MHz/50MHz Current Probe
Martin72:
Hi,
To get an idea why the micsigs are highly interesting, four other currentprobes (pic below, don´t get it why the pic in post function still won´t work).
Testec and the "cheap" R&S doesn´t have degaussing function and worse basic accuracy, lower bandwith.
Rigol and the second R&S do have this like the micsig, also the bandwith and the basic accuracy when you look only on the percent (1%).
Rigol needs an additionally supply.
But the main thing is, look at their prices....
Hope, Dave could do a test soon.
Martin
EEVblog:
--- Quote from: Martin72 on November 25, 2022, 10:55:35 pm ---But the main thing is, look at their prices....
--- End quote ---
That's the kicker. Still I bet a lot of people will say they are expensive.
Martin72:
Yeah because they don´t know what´s behind I guess.
First the bandwith, 50Mhz or 100Mhz are "outstanding", then the measure principle itself.
For decades you could get this only from tektronix, later from hioki.
Prices between 3000...5000.
The first "cheap" one I knew was from Instrance, I´ve bought it 2yrs ago.
User TopQuark got another one from another "brand".
Problem of the instrance was noise ( not too bad but remarkable) and overshoots.
Now the attempt of micsig...
Knowing their CP2100 probes I have faith in it that they won´t perform too badly.
Let´s see if micsig will do it better than instrance.
Martin72:
Anything new here?
jonpaul:
Bonsoir just saw this interesting thread.
Have acquired, used and tested current probes since 1970s, as SMPS, electronic ballast and consulting.
Used them a lot in debugging,'uch more information than just voltage probes.
Our collection is Tektronix, Yokogawa, and Pearson (passive CT)
ranges up to 100A DC and 1000A AC, with Tektronix matching amplifiers.
Usual test with Tektronix PG506 or PG508 into 50 Ohms term to make 1x or 10x loop.
Issues to check
3 db BW, (not 6 db!) rise time,pulse abberation ( overshoot)
DC offset adjustment,degauss effectiveness, DC drift with warmup and temperature
CMRR eg with 500V H bridge wire the fast rise on the probed wire will couple into the currency probe
Easy test probe wire with 0 current, drive wire from HV pulse gen or H bridge,find CMRR of probe.
Cables are reasonable thickness and flexibility on Tek except old P6042.
Any probe requiring power may be specific to an input scheme that's propriétaire.
Scope probe external power sockets are not standardized, eg identical Lemo 4 pin on Tektronix and Yokogawa but 12vs 15 V and different pinout
So, beware the fragile cores, dropping the probe even 100mm on hard surface can cause hairline cracks.
Would reccomend the vendor send Dave a sample for test and eval BEFORE any qty order.
If one is willing, ship a probe to me ( have USA drop) and I can test and write an evaluation.
Just the ramblings of an old retired EE