Author Topic: Micsig Current Probe CP2100A/B Tests and Comparing  (Read 35085 times)

0 Members and 1 Guest are viewing this topic.

Offline Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 5833
  • Country: de
  • Testfield Technician
Micsig Current Probe CP2100A/B Tests and Comparing
« on: May 10, 2020, 08:40:16 pm »
Hi,

https://www.eevblog.com/forum/testgear/350mhz-scope-2020/msg3056674/#msg3056674

I decided to order one, take it to work and hopefully can directly compare to other probes we use.
100Apk, 2.5Mhz Bandwith….sounds nice.

Price is 375€ (Batronix), well, what do we get from others:

Rigol RP1002C for 1886€

Testec TT-CC990 for 1178€

R&S RT-ZC02 for 898€

Siglent CP4070A for 842€

None of them offers the claimed 2.5 Mhz bandwith, some offer more than 100Apk, some less.
But all got one thing in common:
They were unlikely more expensive.
Therefore, the Micsig is very interesting.
The Micsig, like all other I´ve linked didn´t got a degaussing function, like I know from our tektronix probes but also from the lecroy ap011 probe.
At home I got the Owon CP05+ Probe, for appx 190€ incl. VAT.
Well....hm....Ah, forget it, this thing is crap.
Bandwith is claimed by 200khz, in real it would be something around 100khz, very noisy, you could use this thing from 1A up.
Before I got the Hantek CC-65 Probe....God... :palm:

For it´s price, I expect the Micsig will do a better job than my Owon or Hantek, that´s for sure.
But what about to compare it with "state of the art" probes?
What we got on work are Tektronix Probes and Lecroy Probe.
Chances are given, I can compare the Micsig with them, if Time allows it.
So I would do a comparison between CP2100B, CP05+, A6302, A6303, AP011.

My greatest "fear" would be the noise level.
Using tek probes or the ap011 from lecroy, you have a really low noiselevel - Could the CP2100B have it too ?

Let´s find it out... ;)
Missing degaussing function doesn´t concern me much, more important is the noise.

But first of all, I´ll send a offer request to batronix, it would be 3% off I guess.
But hey, taking all what you can get..

Nevertheless, the probe should arrive within the forthcoming week.

Any test suggestions ? Post it here.
DC current is self explaining, so no worries.
AC current I could measure up to 400hz.
Switching power current upt to say 10...20khz, depending on the DUT I could get on work.

To be continued..

« Last Edit: July 23, 2020, 10:07:35 pm by Martin72 »
 
The following users thanked this post: egonotto, thm_w, 2N3055, Elasia

Offline Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 5833
  • Country: de
  • Testfield Technician
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #1 on: May 11, 2020, 10:45:43 am »
Quote
But first of all, I´ll send a offer request to batronix, it would be 3% off I guess.
But hey, taking all what you can get..

No discount at all, but it doesn´t worry me.
When it´s good, then it´s got excellent bang for bucks value(see first post).
If it fails (noise..), I´ll return it and got my money back.
Should arrive in 2 days...
 
The following users thanked this post: egonotto

Offline 2N3055

  • Super Contributor
  • ***
  • Posts: 6631
  • Country: hr
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #2 on: May 11, 2020, 11:23:55 am »
Quote
But first of all, I´ll send a offer request to batronix, it would be 3% off I guess.
But hey, taking all what you can get..

No discount at all, but it doesn´t worry me.
When it´s good, then it´s got excellent bang for bucks value(see first post).
If it fails (noise..), I´ll return it and got my money back.
Should arrive in 2 days...
Looking forward to see how it fares. Thank you for the effort.
For low level testing, few loops of wire and serial 50 Ohm on a signal generator could give interesting signal to see how it behaves down in noise land....
 
The following users thanked this post: Martin72

Offline Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 5833
  • Country: de
  • Testfield Technician
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #3 on: May 11, 2020, 11:28:11 am »
Good idea  :-+

Offline Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 5833
  • Country: de
  • Testfield Technician
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #4 on: May 11, 2020, 07:45:10 pm »
At last, I´ve found some specs for the lecroy ap011 which I want to compare with the micsig - from an old, downloadable catalogue.
(Would upload here because of some interesting articles inside, but it´s too big)
From the spec side, the ap011 looks ( of course ? ) better what accuracy concerns.
But I won´t know the original price for this… 8)
Got the Information from DHL, the probe should arrive on wednesday.
First I will do some tests here at home, like 2N3055 suggest one, together with the owon probe.
After this, I want to take the both to work and trying to find something interesting to measure and compare.



Offline Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 5833
  • Country: de
  • Testfield Technician
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #5 on: May 12, 2020, 05:30:45 pm »
Quote
Got the Information from DHL, the probe should arrive on wednesday.

No….Today.  ;)

First impressions, good building quality, clamp got not a gap between when it´s closed ( like the owon did..), very solid cable, not bad.
Maybe tomorrow I start some measurings or at the forthcoming weekend.

 
The following users thanked this post: 2N3055

Offline TurboTom

  • Super Contributor
  • ***
  • Posts: 1389
  • Country: de
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #6 on: May 12, 2020, 08:57:22 pm »
Have you got an option to evaluate the inductance of the probe? @100A, even a few nanohenries can be a lot. If too high, this may limit the usefulness of the bandwidth at the high current range.
 

Offline Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 5833
  • Country: de
  • Testfield Technician
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #7 on: May 12, 2020, 09:04:35 pm »
Hi, I´m afraid not.
But have a look at the specs.


Offline nctnico

  • Super Contributor
  • ***
  • Posts: 26906
  • Country: nl
    • NCT Developments
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #8 on: May 12, 2020, 09:10:05 pm »
At high currents you can compare a shunt resistor versus the probe. You just need a way to send a large current surge.
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 

Offline Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 5833
  • Country: de
  • Testfield Technician
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #9 on: May 12, 2020, 09:19:19 pm »
Shunts we got a lot, we got also ZES LMG power analyzers with high precision current transformers or internal shunts.
For a surge current, I think I could use our selfbuild inductance meter.
It was designed to measure filter coils nearly to their normal operating current.
We calculating the inductance through the voltage-time area via pulses send from an µC to a IGBT.



Offline Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 5833
  • Country: de
  • Testfield Technician
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #10 on: May 16, 2020, 01:18:22 pm »
First time playing around with it, some simple DC measures..
To get a higher current measure, 4mm banana cable was turning five times through the probe, power supply deliever 3A max.
Really nice: On the siglent you could choose between voltage or ampere in the channel menu, then the probe ratio will also changed from 1:10 etc. to V/A, makes it easy to find the right range.

Noiselevel….
Picture shows the noiselevel at 50mA/Div, for me it´s acceptable and a nice surprise.
Channel 4 shows the noiselevel from the owon probe.... :P
Current was measured by my calibrated brymen bm869s.
At 10A the micisg beeps a warning and the 10A button is flashing, reminding you to change to 100A, nice.
This was the first play around, more to come soon.. ;)
« Last Edit: May 16, 2020, 02:51:25 pm by Martin72 »
 
The following users thanked this post: egonotto, thm_w, 2N3055, maginnovision

Offline maginnovision

  • Super Contributor
  • ***
  • Posts: 1963
  • Country: us
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #11 on: May 16, 2020, 08:29:09 pm »
I was hoping to add some extra probes to the mix but my bnc adapters are VERY noisy. A short across them had more noise than the owon. I suspect the ground is making poor/no connection since they're all safety adapters so you I don't die.

Looks good so far. The owon looks like it'd work much better for a dmm.
 

Offline Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 5833
  • Country: de
  • Testfield Technician
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #12 on: May 16, 2020, 08:56:13 pm »
I think the same about the owon.
Nearly a year ago, I had a thread started about it, doing a schematic from it.
It got a very simple circuit design and the claimed 200khz bandwith couldn´t be right.
Nevertheless, the "Noisetest" was positive, so I will take the micsig to work instead of returning to batronix.
What I will do there is a comparison between lecroy ap011 and, maybe a little unfair, to the good old tek probe.
Next time ac currents will be measured, also I try to measure the bandwith/risetime.
If I get the chance, I´ll do some measures on "real" DUTs ( we are in the market of building power inverters up to 200KVA, 50/60/400hz, switching frequencies between 1.62....20khz)
Tomorrow I´ll try 2N3055´s suggestion.

Edit:

What else I want to try is using a power bank instead supplying the probe  from a USB port, to see if the (actually low) noise level could be further decreased with it.
« Last Edit: May 16, 2020, 10:13:57 pm by Martin72 »
 
The following users thanked this post: egonotto, 2N3055

Offline Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 5833
  • Country: de
  • Testfield Technician
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #13 on: May 17, 2020, 03:53:42 pm »
Did it..

Using the nice bnc-adaptors, I took the internal generator as source and the "load" was the internal 50ohm Input, then turning 10 wires through the clamp.
Below some measurings like risetime, several sinewave frequencies and....If I got bode plot on board, why not use it...So I´ve done it.
The testsetup especially with the 10 windings could not be perfect but at last, I´m surprised about the micsig in a very positive way.
Next step, comparing at work….
Martin

edit must measure the risetime again, wrong channel... :P

risetime added
« Last Edit: May 17, 2020, 04:58:09 pm by Martin72 »
 
The following users thanked this post: Kean, 2N3055, Algasman

Offline 2N3055

  • Super Contributor
  • ***
  • Posts: 6631
  • Country: hr
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #14 on: May 17, 2020, 05:29:27 pm »
That is decent small signal response, and noise is quite low..  Nice work Martin! Looking forward to next episode!
 
The following users thanked this post: Martin72

Offline voltsandjolts

  • Supporter
  • ****
  • Posts: 2300
  • Country: gb
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #15 on: May 17, 2020, 06:25:07 pm »
Micsig has been a pleasant surprise for me too.
I've bought two of their tablet scopes, best portable scope for the money IMHO.
Will be interested in your test results.
 
The following users thanked this post: Martin72

Offline Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 5833
  • Country: de
  • Testfield Technician
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #16 on: May 17, 2020, 07:18:47 pm »
Next week is a short one due public holiday on thursday, so I have three days to "steal" a little bit time and some gear..
Goals are the mentioned comparison between tektronix 50Mhz clamp and the lecroy what accuracy, risetime/bandwith concerns.
Also I want to measure "real" high current and not turning windings through the clamp.
Well, we´ll see what happen.


Online tautech

  • Super Contributor
  • ***
  • Posts: 28377
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #17 on: May 18, 2020, 03:56:05 am »
Martin, in this screenshot there appears some significant current probe propagation delay:



Can you use the channel 1 deskew to see how much delay the MICSIG current probe inserts into the measurements ?
Does it meet the manufacturers propagation delay spec ?
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 5833
  • Country: de
  • Testfield Technician
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #18 on: May 18, 2020, 04:49:00 am »
Hi,

Will do it in the next days.

Martin

Offline Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 5833
  • Country: de
  • Testfield Technician
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #19 on: May 18, 2020, 05:12:47 pm »
Comparing three probes noiselevel at 50mA/div. on the hdo6043a, Tektronix, Lecroy and Micsig.
"Of course" the Micsig is noisier, but not much - Excellent.
@tautech: Probe is at work, will looking for, if the lecroy got the deskew function too.
 
The following users thanked this post: egonotto, 2N3055

Online tautech

  • Super Contributor
  • ***
  • Posts: 28377
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #20 on: May 18, 2020, 07:15:58 pm »
@tautech: Probe is at work, will looking for, if the lecroy got the deskew function too.
All half decent scopes have channel deskew.  :P

SDS5054X SDS5104X with 3 current probes....ch1 probe broken  :(
Ch2 menu showing deskew WRT the ch4 voltage probe.

Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 5833
  • Country: de
  • Testfield Technician
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #21 on: May 18, 2020, 07:29:48 pm »
Quote
All half decent scopes have channel deskew.

Never used this, so I must look at it.  ;)

Offline 2N3055

  • Super Contributor
  • ***
  • Posts: 6631
  • Country: hr
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #22 on: May 18, 2020, 09:27:42 pm »
Quote
All half decent scopes have channel deskew.

Never used this, so I must look at it.  ;)
It's in Channel setup, in preprocessing, on the right..
 
The following users thanked this post: Martin72

Offline Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 5833
  • Country: de
  • Testfield Technician
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #23 on: May 20, 2020, 08:20:31 pm »
Next week is a short one due public holiday on thursday, so I have three days to "steal" a little bit time and some gear..
Goals are the mentioned comparison between tektronix 50Mhz clamp and the lecroy what accuracy, risetime/bandwith concerns.
Also I want to measure "real" high current and not turning windings through the clamp.
Well, we´ll see what happen.

Unfortunately, I got no time at work this week - Except this one..
"Real" DC current measure up 100A.
The tek probe was in use, so I compare it with the lecroy ap011.
"reference" was a shunt, voltage measured by fluke 87V in hires mode, scope was a lecroy hdo6034a.
Source was a 100adc supply in constant current mode.
The cp2100b was every time in it´s tolerance values, also the ap011 - Remarkable: The claimed 15% from 40A up wasn´t reached at any time, worsest deviation was 4.22% at 100A.
Nice..
Hopefully, AC measure and switching current measure I could make next week on work.
I´ve took it back to home, making the measure tautech asked for it.

« Last Edit: May 20, 2020, 08:26:42 pm by Martin72 »
 
The following users thanked this post: thm_w, 2N3055

Offline Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 5833
  • Country: de
  • Testfield Technician
Re: Micsig Current Probe CP2100B Tests and Comparing
« Reply #24 on: May 24, 2020, 03:04:02 pm »
Can you use the channel 1 deskew to see how much delay the MICSIG current probe inserts into the measurements ?
Does it meet the manufacturers propagation delay spec ?

Hi,

Specs were <150ns (10A), <200ns (100A).

Below 2 pics without deskew and with max deskew on ch1.

 
The following users thanked this post: 2N3055