Edit: reading Andy's reply, yeah make sure you're not blasting the input of your scope with the power from the transmitter. Remember that the maximum input derates with increasing frequency.
When triggering on the RF burst, have you tried adjusting the trigger holdoff? Sometimes this works, but you might also not be able to trigger off the modulated signal in your case.
Not sure if this answers your question but the ASV trigger is either in free-run or trigger on power level (only the FFT window).
ASV configures the scope continuously while it takes control of the scope to pull data, so you cannot manually change the horizontal setting on the scope during ASV.
I can't duplicate your setup here at home but what I tried was (using the demo mode signals of the scope):
- start the ASV software and once it runs, click on 'Stop' (the application loves to reset your scope to default)
- enable the I2C training signal (demo 2 is SDATA, demo1 is SCLK)
- to feed the I2C training signals to the digital pod D0(SCL), D1(SDA)
- Disable all other serial channels with the digital menu, threshold 1.4V (the training signal is 2.8 Vpp)
- Serial decode menu set to serial 1, I2C mode, 7 bit address size,
- in the trigger menu set the trigger type to S1(IC2)
- set to trigger on write7 i.e. frame(start:addr7:write:ack:data) with Address set to 0x29 and data 0x01 so you can see the data cycle (20, 40, 60, 80 etc )
(adjust the trigger to whatever is meaningful in your application)
you can now use either the analog channel 1 or 3 to probe the RF signal, although 433 MHz burst on a 500 MHz BW scope... hmja, depends on your modulation type I guess.
- in the ASV software, click on 'Run' and adjust the start/stop frequency and RBW to what you need.
You should be able to at least see the carrier with your scope, not sure what the stop frequency can be set to with the MSOX-3054A