Author Topic: My Philips PM3050 Oscilloscope is giving wrong frequency from cal output. Why?  (Read 350 times)
Offline TulparTopic starter
  • Country: tr
My Philips / Fluke PM3050 Oscilloscope is giving the wrong frequency value from the Cal output. Does anyone know why?
The user manual says that the calibration output should be 1.2V (p-p) 2kHz. When I measure, I see a value of approximately 0.57 ms = 1.75 kHz. I also tested with another oscillator and the oscilloscope can measure 2 kHz as it should. Does anyone have any ideas on how to fix the calibration output or what the problem is?
Offline Wallace Gasiewicz
The Cal output is really just to use as probe calibration.  They are used to calibrate the probes for best square wave. Yours is off more than I would hope for but most of them are not accurate as freq standard, they are usually off quite a bit.  The voltage is usually pretty good but although scopes are good volt meters they are not excellent volt meters,
The following users thanked this post: Tulpar
  • Country: tr
I thought that the calibration output would give an output at the frequency written in the user manual. The square form of the wave seems to be correct, I guess there is no problem then. Thank you for your help.
Online tggzzz
The cal out signal would be just as effective at 1kHz or 5kHz. Its sole purpose is to enable you to tweak the LF response of *10 "high" impedance probes.