Basically, I got a set of HP 8510C for free recently, but it didn't come with any original cable.

It needs some non universal cable that costs more than 400$ on ebay, I probably need to figure out a way to make them myself.



08510-60101 for connection between VNA and display, both male 37 D-SUB on each side.


08510-60102 for IF signal between S-parameter test set, 24W7 D-SUB with coaxial contact, one male and female on the other.


I ordered the right connector already, now the problem is to findout the correct order. Can anyone with 8510 check their pinout of these two with a multimeter?

Thanks for any related clues. :-+
« Last Edit: October 13, 2024, 06:27:23 pm by BillyGo »
 

Offline squadchannel

  • Frequent Contributor
  • **
  • Posts: 302
  • Country: jp
  • deepl translate user
Re: Needs help of HP 8510 cable DIY
« Reply #1 on: October 07, 2024, 01:25:05 am »
I thought I could find it in the service manual, but there was no schematic.

 :o
https://www.glkinst.com/cables/hp_cables.htm#8510
 

Offline BillyGoTopic starter

  • Newbie
  • Posts: 9
  • Country: ie
Re: Needs help of HP 8510 cable DIY
« Reply #2 on: October 07, 2024, 02:44:02 pm »
I thought I could find it in the service manual, but there was no schematic.

 :o
https://www.glkinst.com/cables/hp_cables.htm#8510

The block diagram in the service manual was too basic.


This guy obviously doesn't provide the best offer  |O
 

Offline bergenhut

  • Newbie
  • Posts: 6
  • Country: au
Re: Needs help of HP 8510 cable DIY
« Reply #3 on: October 14, 2024, 10:42:53 am »
Hi,
Here are some pics of the 2 cables for the HP 8510 system in question
They are 1 to 1
Both cables are fully sheilded, with the sheild at both ends connected to the metal cases and Not to any pins.
The 08510-60101 cable- Pin 37 is NOT connected at both ends
 
The following users thanked this post: BillyGo

Offline BillyGoTopic starter

  • Newbie
  • Posts: 9
  • Country: ie
Re: Needs help of HP 8510 cable DIY
« Reply #4 on: October 21, 2024, 12:06:01 am »
Thank you so much!!

Looks like 08510-60101 can be replaced by any normal DB37 cable!

I couldn't identify whether the signal wires of 08510-60102 were twisted pairs. From the picture they were the same as the none twisted cable inside 08510-60101. I might try with some normal 23AWG cables for my DIY first.
 