I looked around on the forum for a dedicated 80 series thread but was unable to find one so for now I will leave this here.
Recently I acquired a Fluke 83 which is noted earlier in this thread and I have a question in regards to the continuity piezo which from my tests is activated at 70 ohms and then deactivated once the reading rises above 196 ohms when set to auto or low range, I am yet to document the continuity thresholds in other ranges which do vary according to the range selection.
Everything on this particular meter seems to be working fine but when in resistance mode and the min-max button is selected the piezo will permanently activate in either min, max or average until the probes are shorted, so the continuity mode is inverted. On a couple of other meters that I have such as the Fluke 187, 189 and an old Metex M-3650CR whilst in continuity mode if the min-max is set to min then the continuity piezo will be off until a momentary short of the probes is applied which will permanently latch the piezo until the unit is manually reset, this is a handy feature for the type of work that I am involved with although I suspect that most users would never operate the meter in this manner.
Other meters that cannot latch the piezo whist in continuity mode with minimum selected and a momentary contact applied are UT-70D, 83, 117 and U1272A, these are the only meters that I have tested, I don't think the 179 did it either but it's not at hand to confirm.
Years ago I made up a little box with a latching relay, 9 volt battery and a piezo buzzer for testing as one example that a gate controller received a momentary contact from an access control system, this was handy whereby we would not have to walk out to the gate to visually inspect a meter and could hear the buzzer from a distance to know that the unit was triggered, we also used it for rudimentary cable testing on occasion but now have dedicated equipment for these applications, the little box went missing so I think that one of the boys left it somewhere and we have been using certain meters to do the same thing.
So in summary if anyone else has a Fluke 83 can you put it in ohms mode and press min-max and let me know if your continuity buzzer activates immediately with the probes open even though the continuity symbol is not displayed, mine does and it seems a bit strange.
Many Thanks
Muttley