I spent a few hours on the beast this evening... For anyone who wants to follow along at home, the service manual is in KO4BB's archive:
http://www.ko4bb.com/manuals/download.php?file=06_Misc_Test_Equipment/Gould/Gould_DSO400.pdf http://www.ko4bb.com/manuals/index.php?dir=06_Misc_Test_Equipment/GouldFault symptoms:
- CAL fails CCD offset cal
- Calibration Hardware Check function reveals that none of the cal hardware is functioning. Gain slope is horizontal (should be a ~45 degree slope) and offset slope is also horizontal (permissible but not really ideal).
- 1V calibrator output on the panel was as dead as a doornail, flat zero output
- Traces can't be moved up/down the screen
First visual check revealed that the CMOS battery (a Varta Mempac NiCd) on the CPU board had leaked and muckied up the connectors and eaten a few traces on the CPU board. Those traces were repaired. Interestingly there are a few traces under the battery which seem to be routed through, then end abruptly! I thought I had three broken traces, when in fact I only had one... (I hope?)
On to Rule Zero, always check the rails -- this revealed that the 4051 which serves as an offset DAC for the preamp chip had no +5V supply. Checking further back, the +12V supply was gone too, along with the input to the +12V regulator in the Preamp/Trigger section. This was traced back to a smoked 2R2 resistor, R12. Replaced that, rails came back up. Traces are still stuck doggedly in the middle.
Further checking revealed there was no VREF+ input to U307 and U314 (two DAC08HPs). These are fed with +5.2V via a 3k9 (one per DAC), and I'm seeing 0V at the VREF+ pin of both. Surely that's not right?
Or do these DACs load down the VREF input?
I must confess I'm not that clear on the how multiplying DACs work... most of my experience is with the more common voltage-output type.

There's also another fault - the power rails seem to have a voltage spike (more accurately a short-duty square-wave, about 200ms high every second) which is causing the screen and FP LEDs to change in brightness. To try and rule out a problem on the AC side of the power supply, I hooked up a 12V supply to the input on the back. A few seconds later, during the self-calibration phase, the sodding 2R2 resistor blew again!

This after an hour or two of running fine (ish) off the mains!
I swear this machine is possessed...