Fluke 117 current shunt burnt
« on: March 03, 2019, 09:09:54 pm »
I got a Fluke 117 to repair. She/he has decided to improve Fluke design and changed the 11A fast fuse with a "better" one, 32A. Now the current shunt R3 is burnt and split in two parts. 32A fuse is fine. Current function fail but the others appear to be fine. Has somebody idea what value is R3?
Thanks for any tips

https://flic.kr/p/9JUe3P
 

Offline killcode

  • Contributor
  • Posts: 13
  • Country: au
Re: Fluke 117 current shunt burnt
« Reply #1 on: March 04, 2019, 01:19:17 am »
You could feed a small known voltage (Id start with a few mV) across the shunt location and use that to figure out what the shunt should be based off what the meter reads.
Assuming the nothing past the shunt is damaged.
 
The following users thanked this post: danno_cj

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Fluke 117 current shunt burnt
« Reply #2 on: March 04, 2019, 05:57:14 am »
The Fluke 17B, a different meter obviously, schematic shows its current shunt as 0.01 ohm.

As a guess, it is likely in the 0.01 ohm ballpark area.
« Last Edit: March 04, 2019, 06:03:01 am by retiredcaps »
 
The following users thanked this post: danno_cj

Offline Gyro

  • Super Contributor
  • ***
  • Posts: 9410
  • Country: gb
Re: Fluke 117 current shunt burnt
« Reply #3 on: March 04, 2019, 10:40:32 am »
With the shunt gone, you ought to be able to see a reading with a voltage source in the order of mV to 100mV across the 10A terminals - if not then I would suspect other damage to the meter. It wasn't expecting to see full voltage of whatever high current source was  being measured, you can protect the lower current ranges with inverse parallel diodes and the small fuse but I doubt they would be present or effective on the 10A range.

If you can get some sort of reading on the 10A range, then you ought to be able to calculate the shunt value. Eg. It it displays 10A for a 100mV input then the shunt resistor would be 0.01R.
« Last Edit: March 04, 2019, 10:43:00 am by Gyro »
Best Regards, Chris
 
The following users thanked this post: The Soulman, danno_cj

Online Fungus

  • Super Contributor
  • ***
  • Posts: 16562
  • Country: 00
Re: Fluke 117 current shunt burnt
« Reply #4 on: March 04, 2019, 11:17:15 am »
Look in the manual for "burden voltage". It's basically the shunt resistance disguised by Ohm's law.

As a guess, it is likely in the 0.01 ohm ballpark area.

Yep.
 

Offline mzacharias

  • Frequent Contributor
  • **
  • Posts: 709
  • Country: us
Re: Fluke 117 current shunt burnt
« Reply #5 on: March 04, 2019, 12:20:16 pm »
Are you sure it didn't burn out a foil run instead? The high-current shunt would be pretty hard to make fail in that way.
 

Offline ModemHead

  • Frequent Contributor
  • **
  • Posts: 731
  • Country: us
  • No user-serviceable parts inside.
    • Mr. ModemHead
Re: Fluke 117 current shunt burnt
« Reply #6 on: March 04, 2019, 12:53:44 pm »
Are you sure it didn't burn out a foil run instead? The high-current shunt would be pretty hard to make fail in that way.
Yes... would love to see a picture of this failure.
 

Offline 0xdeadbeef

  • Super Contributor
  • ***
  • Posts: 1570
  • Country: de
Re: Fluke 117 current shunt burnt
« Reply #7 on: March 04, 2019, 12:59:41 pm »
Also there's usually a bridge rectifier (and some diodes) to protect the shunt (and ADC) which vaporizes before the shunt dies.
Trying is the first step towards failure - Homer J. Simpson
 

Offline danno_cjTopic starter

  • Newbie
  • Posts: 4
  • Country: ro
Re: Fluke 117 current shunt burnt
« Reply #8 on: March 04, 2019, 04:31:21 pm »
Connecting my power supply to Com and 10A I got reading. At 10mV is around 1A and at 100mV is around 10A.  At 10mV (the lowest voltage of my power supply) is not stable at all but at 100mV it is stay better. It jump up and down because I think my power supply is not stable at low voltage and/or is noisy. So, the R3 value is 0.01R. No other damage appear to be on this PCB side. I don`t know what the owner has done but is safe to say that his load was under 32A.
Here is the picture:
 

Online MarkF

  • Super Contributor
  • ***
  • Posts: 2523
  • Country: us
Re: Fluke 117 current shunt burnt
« Reply #9 on: March 04, 2019, 05:12:40 pm »
I got a Fluke 117 to repair. She/he has decided to improve Fluke design and changed the 11A fast fuse with a "better" one, 32A. Now the current shunt R3 is burnt and split in two parts. 32A fuse is fine. Current function fail but the others appear to be fine. Has somebody idea what value is R3?

That is not a "design improvement".  :scared:
That is "removing the input protection"!
Hence, the burnt front end.

Replace the shunt and replace the fuse with the correct value!!!
« Last Edit: March 04, 2019, 05:14:48 pm by MarkF »
 

Offline ModemHead

  • Frequent Contributor
  • **
  • Posts: 731
  • Country: us
  • No user-serviceable parts inside.
    • Mr. ModemHead
Re: Fluke 117 current shunt burnt
« Reply #10 on: March 04, 2019, 05:19:24 pm »
Here is the picture:
Holy crap!  :scared:

BTW, I measured an (intact) shunt, and it is indeed 0.01Ω.
 

Offline floobydust

  • Super Contributor
  • ***
  • Posts: 6927
  • Country: ca
Re: Fluke 117 current shunt burnt
« Reply #11 on: March 04, 2019, 05:45:30 pm »
Fluke 83, 85, 87 (R6), Fluke 88
10 milliohm 1W 0.1% 100ppm Fluke P/N 807305  (old) Mfgr. P/N R-3269010

Fluke 78 (R11), Fluke 29/79 (R6)
10 milliohm 1W 0.25% 100ppm Fluke P/N 877076

Fluke 77
5 milliohm 0.5W 1% Fluke P/N 720415

Fluke 27 (R23)
5 milliohm 0.5W 0.5% Fluke P/N 655423

Fluke 45, 867B, 863
10 milliohm 1W 5% 100ppm 4-wire Fluke P/N 820845
 
The following users thanked this post: lowimpedance, Marco1971

Offline floobydust

  • Super Contributor
  • ***
  • Posts: 6927
  • Country: ca
Re: Fluke 117 current shunt burnt
« Reply #12 on: March 04, 2019, 06:12:36 pm »
Here is the picture:


It looks like somebody connected it to a car battery, low voltage so no arcing or burns on the pcb.  The test leads must have smoked too.
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Fluke 117 current shunt burnt
« Reply #13 on: March 05, 2019, 04:10:48 am »
The solder leg to RV1 looks like it has a solder crack.
 

Offline Electro Detective

  • Super Contributor
  • ***
  • Posts: 2715
  • Country: au
Re: Fluke 117 current shunt burnt
« Reply #14 on: March 05, 2019, 08:40:54 am »

Luckily it wasn't oops-graded to a 100 amp T-Fuse   :scared:

If the 117 can be fixed, the owner hack/reverse engineer needs to seriously consider a decent cheap AC-DC clamp meter
to keep the 117 company and out of harms way...

and yeah, how did the barbequed leads fare?   :o
 

Offline Cnoob

  • Regular Contributor
  • *
  • Posts: 132
  • Country: gb
Re: Fluke 117 current shunt burnt
« Reply #15 on: March 05, 2019, 09:57:07 am »
If I want to measure any high current above 8 Amps or so I use an external shunt. I brought a 30 Amp one from Ebay
[urlhttps://www.ebay.co.uk/itm/10A-50A-75mV-Shunt-Resistor-for-DC-Current-Meter-Amp-Analog-Panel-Ammeter/352365629434?hash=item520aa0cffa:m:mylcWBpG3S8CTHdJm5diT6Q:rk:79:pf:0][/url]

I don't like when manufactures state you can measure up to 20 Amps for X amount of seconds it encourages misuse and even at 10 Amps
 they state let the meter cool down. 
 

Offline Electro Detective

  • Super Contributor
  • ***
  • Posts: 2715
  • Country: au
Re: Fluke 117 current shunt burnt
« Reply #16 on: March 05, 2019, 09:31:59 pm »
If I want to measure any high current above 8 Amps or so I use an external shunt. I brought a 30 Amp one from Ebay
[urlhttps://www.ebay.co.uk/itm/10A-50A-75mV-Shunt-Resistor-for-DC-Current-Meter-Amp-Analog-Panel-Ammeter/352365629434?hash=item520aa0cffa:m:mylcWBpG3S8CTHdJm5diT6Q:rk:79:pf:0][/url]

I don't like when manufactures state you can measure up to 20 Amps for X amount of seconds it encourages misuse and even at 10 Amps

 they state let the meter cool down.



lol especially when the internal fuse is 11 or 15 amps (and EXPENSIVE)
and no mention of ambient temperature vs 'X seconds' derating stuff,
or lead length, wire gauge, contact/connector resistance etc

or how to tell when it's cooled down enough for another shot   :-//

I wonder (not) what would happen to some of the meters out there with UNFUSED 10 and 20 amp sockets  :scared:

If a clamp won't do the job and I need to break into a circuit, I use fused leads (or a knockup bodge job with croc clips  :palm: ) with any value below the meter's internal fuse

 