Author Topic: NEW Keysight HD3  (Read 125573 times)
Offline simone.pignatti
Re: NEW Keysight HD3
« Reply #375 on: September 04, 2024, 03:29:34 pm »
Also, it only has a 40MHz channel bandwidth selection, not 20MHz. Why?
That's twice the bandwidth, bigger is better.
But not comparable anymore with all the 20 MHz bandwidth PSU noise measurements out there.
It seems that even that 40 MHz is an digital filter. Not an analog filter in actual front end. Actual order of lowpass filter is not specified.
Each analog input channel has a 40 MHz BW Limit option that lets you disable or
enable a bandwidth limit (using a digital filter). Limiting the channel bandwidth
lets you remove unwanted high frequency noise from the waveform and increase
the number of bits of vertical resolution.
Definitely any measurement out there that was specified with 20MHz will not come out the same on this scope.
HD3 has hardware bandwidth limits 5, 10, 20, 50, 100, 200, 350, 500 MHz, 1 GHz (selectable)
At the moment is a general setting for all channels, on a next FW upgrade will be available for each single channel.
« Last Edit: September 04, 2024, 03:34:49 pm by simone.pignatti »
Technical Support
Offline DaneLaw
« Reply #376 on: September 04, 2024, 03:47:06 pm »
Options - Licenses. [#369]
Offline tszaboo
« Reply #377 on: September 04, 2024, 03:57:49 pm »
I think the joke went right over your head, even though I put a clarification the next line...
Offline tggzzz
« Reply #378 on: September 04, 2024, 04:18:57 pm »
 (Even if I can't buy it because Keysight refuse sell it to people like me).
Why you want to buy directly from Keysight?  :-//
You can buy from a distributor (which can offer you support in case you need).
Search for "Keysight support" here and elsewhere, e.g. one current thread at https://groups.io/g/HP-Agilent-Keysight-equipment/message/146632
I know about, but you said that they will refuse to sell. If you really want something from KS you can have it through an distributor like  Batterfly for ex.  @simone.pignatti . You can request support to them, they will open a case to KS as a business and get the support. KS cannot refuse them.
Will the specific distributor still be in business when I need them?
In the UK distributors are Mouser, Farnell, RS and a few other companies I haven't heard of. (RS sells Keysight scopes that are £990k+VAT) Do you think those distributors will give you effective support? RS certainly fails on support for their basic services such as calibrating a voltage source - their staff can't even comprehend that it is necessary to state the uncertainty in their measurements!
Keysight appears not to supply businesses unless they also have a website. If they will make that kind of idiotic equivalence, then what will they be like in the future?
Offline CRTbrain
« Reply #379 on: September 04, 2024, 04:30:27 pm »
Can anyone confirm the HD3 uses a 14-bit ADC, or is it really the same 10-bit ADC used in MXR/UXR locked into 14-bit high res mode?   Historically, Keysight has had more technology info about new ADCs and ICs in new scopes.  I don't see anything like this.
The following users thanked this post: Someone, egonotto
Offline Keysight DanielBogdanoff
  • ALL THE SCOPES!
    • Keysight Scopes YouTube channel
« Reply #380 on: September 04, 2024, 04:41:40 pm »
yes, confirmed it's a 14-bit
Offline maxwell3e10
« Reply #381 on: September 04, 2024, 04:43:25 pm »
First noise measurement.
50R 500uV 1ms 1M point 100MS/s 1GHz
I looked briefly at the file, there are a couple of problems. First it has only 62500 points in it, not 1M.  Perhaps some decimation happens in saving. Also 100 MS/s rate is not sufficient to get accurate noise spectrum up to 1 GHz, so it becomes aliased. You can reduce the time/div until the sample rate is maximum or see what happens if one turns on the bandwidth limiting filter.
The following users thanked this post: EEVblog, egonotto
Offline nctnico
« Reply #382 on: September 04, 2024, 04:52:07 pm »
Can anyone confirm the HD3 uses a 14-bit ADC, or is it really the same 10-bit ADC used in MXR/UXR locked into 14-bit high res mode?   Historically, Keysight has had more technology info about new ADCs and ICs in new scopes.  I don't see anything like this.    (Attachment Link)
It could make sense Keysight uses the 10 bit ADC from the UXR and oversamples to get more bits. 256Gs/s / 80 = 3.2Gs/s. Oversampling 80 times gives over 3 bits on top of the 10 bit. Make it 14 because many buildings in the US don't have a 13th floor either.
The following users thanked this post: egonotto, skander36
Offline jusaca
« Reply #383 on: September 04, 2024, 05:04:42 pm »
Ohhh my bad, I did not get that line  :-+
The following users thanked this post: tszaboo
Offline Zucca
  • EE meid in Itali
« Reply #384 on: September 04, 2024, 05:59:33 pm »
Can't love what you don't
The following users thanked this post: Howardlong
Offline 2N3055
  • Country: hr
« Reply #385 on: September 04, 2024, 06:10:28 pm »
Simone,
I am quoting User manual.
What you are saying is actual Global BW limit that is also a digital filter, specifically a Hires mode where BW limit is specified in a same way R&S does.
No analog BW limit filters of any kind are mentioned anywhere. I presume channels have 1GHz AA filters before ADC but that seems to be it..
If said digital filters per channel (that are coming) are made to mimic classic 20MHz filters by it's cutoff characteristics than it is fine. Apart from fact that signal that enters ADC  is full BW all the time, so on signals where there is significant amount of energy above 20MHz we cannot use filter to make ADCs life easier...
Best,
Siniša
Dr. Richard W. Hamming
Offline electr_peter
  • Country: lt
« Reply #386 on: September 04, 2024, 06:18:37 pm »
Dave can test what exactly the effect is of 40MHz channel filter and 5/10/20/50/100/200/350/500MHz/1GHz global filter. Requires only high BW noise source and FFT. Filter drop-off could be brick-wall or smooth :-//
Offline Anthocyanina
  • The Sara
« Reply #387 on: September 04, 2024, 06:21:29 pm »
ok, so, this weird space next to the analog inputs. i initially thought it would be for the AWG and that the originally leaked photos were of a model without AWG, but since it's been confirmed the AWG is on the back, the space next to channel 4 has to be for something, right? i mean, it could be just a design choice, since not many of the front panel segments are aligned with much else, like the cavity for the channels not being aligned with the display at all or the control knobs, or it not being symmetrical, the lip at the bottom in particular not being equally wide on both sides is weird... anyway, they included MSO in the model number, so that together with the space there is making me wonder if they are planning an MDO model and that space there is for an SA input? 🤔🤔🤔
« Last Edit: September 04, 2024, 06:25:58 pm by Anthocyanina »
Offline tszaboo
« Reply #388 on: September 04, 2024, 06:31:41 pm »
Usually that was the space for external trigger.
The following users thanked this post: HighVoltage
Offline Zucca
  • EE meid in Itali
« Reply #389 on: September 04, 2024, 06:32:31 pm »
or a fifth channel?  :o
  • The Sara
« Reply #390 on: September 04, 2024, 06:37:21 pm »
but that's also in the back!
  • Country: lt
« Reply #391 on: September 04, 2024, 06:50:26 pm »
That space is obviously reserved for credit card reader - SW license activation would be way quicker :D
The following users thanked this post: fcb, egonotto, Zucca, MadTux, jusaca, abeyer, Anthocyanina
Offline Zucca
  • EE meid in Itali
« Reply #392 on: September 04, 2024, 07:00:42 pm »
Offline nctnico
« Reply #393 on: September 04, 2024, 07:03:02 pm »
Does it support Ali-pay, Apple pay and Google pay? Tap your phone to have 10 minutes of protocol decoding.
The following users thanked this post: egonotto, MadTux, tooki, KungFuJosh, Martin72
Offline Martin72
« Reply #394 on: September 04, 2024, 07:27:01 pm »
To be honest, I don't understand why people are so bothered by the free space. ;)
Offline tv84
  • Country: pt
« Reply #395 on: September 04, 2024, 07:37:21 pm »
I don't know if it's the lighting or the color, but the plastic looks cheap. And that space adds to it.
Offline mikeselectricstuff
    • Mike's Electric Stuff
« Reply #396 on: September 04, 2024, 07:43:54 pm »
It does look like something is missing - very annoying for the wavegen to be on the back & not  here
Day Job: Mostly LEDs
The following users thanked this post: nctnico, egonotto, HighVoltage
Online KungFuJosh
  • TEAS is real.
« Reply #397 on: September 04, 2024, 08:13:28 pm »
True, but can you believe how far oscilloscopes have come?? It's only 2024, and already we have a $22,000 scope with a huge 10-inch screen! 🤯🤯🤯
Best Continuity Tester Ever
The following users thanked this post: egonotto, edavid
Offline Martin72
« Reply #398 on: September 04, 2024, 08:15:49 pm »
I have to admit that the look of the whole device is rather unfortunate.
And the fact that the screen is rather small, considering the price range, is even more irritating.
But perhaps there's a deeper meaning behind it that we can't yet grasp.
Oh yes, I just love the multiple measurement windows.
Even the small Rigol DHO800 has them, albeit not perfectly, but still.
I can only dream of something like that. ;)
The following users thanked this post: egonotto, MadTux, KungFuJosh
Offline fcb
    • Electron Plus
« Reply #399 on: September 04, 2024, 08:29:37 pm »
There's something unfinished about the front panel design.  Like it's been rushed out, or the tooling modified at the last minute or a major rejig of the PCB.
My guess is that it was a 6 or possibly 8 channel design that got it's wings clipped late on in development for some reason (thermal/cost/performance or they needed more physical width per channel) with the missing input channels would have extended all the way to the right.  The button's have then been spaced more to take out the spare real-estate.  The thick black bezel around the display also makes me think they had bigger screen initially.  When I look at the molding design and button layout of an MSOX4104G compared to this HD3 I wonder what the real story was behind the development.  Anyone torn one down yet?
The following users thanked this post: shabaz