Author Topic: New Rigol DS1054Z oscilloscope  (Read 2232221 times)
Offline Karel
Re: New Rigol DS1054Z oscilloscope
« Reply #4450 on: June 25, 2019, 02:55:02 pm »
No need to fill in any forms, here you can download the firmware directly from the Chinese Rigol website:
https://rigol.com/Support/SoftDownload/3
Offline MarkF
« Reply #4451 on: June 25, 2019, 04:09:45 pm »
Or the Rigol North America site
http://beyondmeasure.rigoltech.com/acton/fs/blocks/showLandingPage/a/1579/p/p-0019/t/page/fm/0
Offline rolycatTopic starter
« Reply #4452 on: June 28, 2019, 02:43:41 am »
I have updated the first post in this thread to reflect the latest firmware version, and the fact that new scopes come with all options unlocked.
I have assumed that the 'hack' is still required to enable 100MHz bandwidth on scopes such as the DS1054Z.
The following users thanked this post: Gandalf_Sr
Offline t_i_t_o
  • Country: bg
« Reply #4453 on: August 01, 2019, 07:38:31 pm »
Continuing my test https://www.eevblog.com/forum/testgear/new-rigol-ds1054z-oscilloscope/msg1109785/#msg1109785 with the same source signal i.e. the 100MHz SDRAM clock of the STM32F769 discovery board, now testing with a 500MHz 8pF Tek probe, it turns out my lovely Rigol has over than 300MHz bandwith (1.3 ns rise time as seen) :)
Offline RoGeorge
« Reply #4454 on: August 01, 2019, 10:44:35 pm »
No, DS1054Z does not have 300MHz bandwidth.
That very small rise time is an artifact created by the DSP post-processing, DSP required for drawing on the screen a continuous line starting from only a few samples, when sinx/x is on and vector mode on.
Offline David Hess
  • DavidH
« Reply #4455 on: August 01, 2019, 11:02:06 pm »
It also varies with signal level and volts/div setting which indicates that something uncouth is going on.
Offline MarkF
« Reply #4456 on: August 01, 2019, 11:37:47 pm »
Some where here (I think in the hugh Rigol thread), the DS1054Z, hacked to 100 MHz, has a measured -3 db bandwidth of 160 MHz.
Note- How about a screen snapshot from the scope instead of a hugh picture from a camera....
« Last Edit: August 01, 2019, 11:41:27 pm by MarkF »
Offline RoGeorge
« Reply #4457 on: August 02, 2019, 03:35:42 am »
I think the display also varies with the spectrum of the incoming signal.  Didn't properly tested yet if the spectrum outside 100 MHz is the real cause, but I noticed that a 30 MHz square wave with a 10ns rise/fall time is displayed reasonably well, while the same 30 MHz and same voltage level from a TTL output (probably a 2ns fall/rise time, not specified in the generator's specs) is very distorted in various ways depending on the on/off status of the sinx/x mode and vector/dots mode.  The slow edge signal was taken from the signal output of a DG4102 DDS generator, while the fast edge signal was taken from the sync output of the same generator which sync output is described as TTL compatible.
I didn't double checked with an analog oscilloscope if the two signals were really the same square wave with the only difference in edges' speed, but I'm almost sure they were the same, because the artifacts I've seen were nothing like normal ringings or reflections.  They were weird and unnatural looking shapes that I've never seen before, clearly visible only when the fall/rise edges were << 10ns (for a 30 MHz square wave, 50%, TTL level, and the oscilloscope sampling rate at 250MSa/s).
Here's an example with sinx/x off, dots mode, run mode, for the same 30 MHz square signal except the second one has sharper edges:
Please note that we see so many dots because the oscilloscope is in run mode and many screens are overlapped, while the generator and the oscilloscope does not have the internal clocks synchronized.  The actual number of samples for a single trace is very small:
Same 30 MHz slow vs fast edges, this time with sin(x)/x on:
Maybe I'll redo the test in a more controlled way and take some snapshots.
« Last Edit: August 02, 2019, 03:55:03 am by RoGeorge »
The following users thanked this post: cjs
Offline Fungus
« Reply #4458 on: August 02, 2019, 09:57:09 am »
I think the problem is due to not zooming to maximum zoom. The DS1054Z does all calculations using "on-screen" pixels so for most accuracy in things like rise times you need to zoom in on the area of interest.
Offline David Hess
  • DavidH
« Reply #4459 on: August 03, 2019, 10:59:34 am »
That has an effect on accuracy but reports of bandwidth vary over more than a 2:1 range.  Some are more than 3:1.  The DSOs I have used have no problem making accurate edge measurements over a 10:1 range in displayed signal amplitudes.  The shape of the waveform should not vary with amplitude.
Like we have discussed in the past, I suspect one of the amplifier stages suffers from slew rate limiting causing saturation or cutoff which also explains the odd 10 nanosecond recovery time under certain conditions.
Offline JohnPen
« Reply #4460 on: August 04, 2019, 09:27:07 am »
Out of curiosity I rechecked the bandwidth of my hacked DS1054Z  over an extended range.  The -3db point is between 130-140 Mhz.  At 200 Mhz I measured -4.8 dB.  250Mhz  was -7.1 dB and 300 Mhz was -9.3 dB.    The signal source was accurate at -7.7 dB at 10M and within +- 0.1 dB over the frequency range measured.  Directly connected to the scope with a through 50 ohm termination.   I also checked the triggering which worked fine to 500 Mhz.  The display frequency measurement was by then jumping around all over the place around 500 Mhz.  Above 100 Mhz don't expect reliable measurements on your own scope unless you can calibrate it yourself and it may well vary with the vertical sensitivity settings.  However you can certainly view signals up to 300 Mhz  with some confidence that they are real.  The triggering, as has been mentioned before,  far exceeds the usable viewing bandwidth of the scope.
Online 2N3055
  • Country: hr
« Reply #4461 on: August 04, 2019, 10:16:03 am »
-3dB point is between 130 and 140 MHz. Period. Measured with frequency sweep from a siggen, 50 OHm pass trough terminator scope side .
The only way.
Outrageous claims are coming from people that either don't understand -3dB part , or people who keep calculating bandwidth from impulse response, ignoring the fact that DS1000Z input freq response has an odd shape that favors edge sharpness and not nice frequency response..
Like JohnPen very nicely wrote (my measurements are very similar to his), you can use to see if your 250MHz clock works without problem.
That doesn't make it 250MHz scope. -3dB point does...
Dr. Richard W. Hamming
Offline TurboTom
« Reply #4462 on: August 04, 2019, 11:52:27 am »
Actually, that doesn't reflect the complete situation: The DS1000Z has two physical, relay-switched sensitivity ranges in the input voltage divider. Further sensitivity adjustments are arranged via the internal digital gain modules in the HMCAD1511 ADC.
As long as the high sensitivity range is selected (this means a less than 330mV/div true input sensitivity), the mentioned figures are correct. But if you change the sensitivity to 335mV/div or above and thus select the low gain input configuration, the 3dB bandwidth increases to somewhere around 220MHz. Of course, this test requires a generator that's capable of some decent output amplitude since the use of a terminator at the scope input is mandatory at these frequencies.
If this behavior is something that's useful in any way may be discussed controversely, I prefer a scope with an input configuration that wouldn't (substantially) change bandwidth when selecting a different sensitivity, especially if it isn't documented anywhere. Anyway, I guess it is what it is and the user's the one who's got to make best use of it. If it helps in some situations -- good, if it doesn't never mind ;) .
I hope this helps to enlighten the situation a little.
Cheers,
Thomas
Online xrunner
  • hp>Agilent>Keysight>???
« Reply #4463 on: August 04, 2019, 12:38:02 pm »
The only way.
Here's my data. I measured 130 MHz for the -3 dB point.
https://www.eevblog.com/forum/beginners/rogol-1054z-actual-bandwidth/msg1380072/#msg1380072
Offline David Hess
  • DavidH
« Reply #4464 on: August 05, 2019, 01:52:31 am »
Besides the inconsistent reports, three things make me think that the bandwidth varies with amplitude:
1. The DS1054Z only has *one* switchable input attenuator which so the input buffer has to operate over a range 10 times larger than many oscilloscopes.  This means the input slew rate is 10 times higher for a given frequency.
2. Per Dave's reverse engineering, the differential amplifier uses 3904s which are only marginally fast enough for a 100 MHz front end even at a lower signal level.  The equalization networks at the emitters show that the designers were pushing the transistors to their limits.
3. The screenshot below shows a non-linearity in the transient response which should not exist.  If this is a 100 MHz 3.5 nanosecond oscilloscope, then at most there should be signs of recovery rather than that flat slope for more than 5 nanoseconds.  To me that looks like a transistor was driven into saturation or cutoff.  The preshoot at 1 GS/s is also questionable.
What that all adds up to is that I suspect large signals applied to the input buffer result in slew rate limiting and besides the transient response issue, this results in a variation of bandwidth with signal level.
Offline TurboTom
« Reply #4465 on: August 05, 2019, 01:43:05 pm »
David -
your assumption regarding the input amplifier getting close to saturation at higher input levels (or encountering problems because of its limited slew rate) sounds reasonable, but it doesn't actually appear to be the case. I just re-checked the bandwidth in the high sensitivity range of my DS1000Z at a setting of 5mV/div with an input signal level of round about 40mVpp.
If your assumption would be correct, I'ld expect a noticably higher bandwidth than with the scope set at 330mV/div, just before the relay switches in the additional analog divider, and fed with an input signal of 1.5Vpp (that's the maximum my SDG6000X will provide with the additional 50 ohms terminator at up to 350MHz). Actually, the bandwidth even appears to be a little higher at the higher input level.
The big difference is when the input sensitivity is changed to 335mV, enabling the divider. Risking to be boring, may I ask to please have a look at the attached screenshots, their file names should pretty well explain what you're seeing there (LR = Low Range, HR = High Range of the relay-operated input divider). I checked the output level of the SDG6000X to be constant within 0.1dB over the frequency range (HP437B + HP8482A).
Actually, I've got the impression that Rigol over-compensated the parasitic capacitances of the additional voltage divider in the high input range, thus emphasizing the high frequencies and causing the HF rolloff to appear slower than it actually is. It will be quite difficult to check this in an open DS1000Z frontend since any probing in this area will affect the performance of the circuitry. It's probably enough to know about this issue and keep it in mind when using the scope. Anyway, the owners buy a 50MHz scope, convince it to believe it's a 100MHz version and then want to use it up to 150+MHz? Sounds weird...  ;)
Cheers,
Thomas
« Last Edit: August 05, 2019, 01:45:07 pm by TurboTom »
Offline Karel
« Reply #4466 on: August 05, 2019, 03:23:03 pm »
Don't you think the DS1054Z and the DS1104Z share the same hardware and that, by using a software key the artificially limited bandwidth can be upgraded?
Offline ebastler
« Reply #4467 on: August 05, 2019, 04:26:13 pm »
I understood TurboTom's comment as just stating that us DS1054Z owners are a greedy bunch.  ;)
Buy 50 MHz, cheat your way into 100 MHz, then hope to get 150 MHz and beyond... 8)
Offline tv84
  • Country: pt
« Reply #4468 on: August 05, 2019, 04:33:29 pm »
Not "hope", "expect".   ;)
« Reply #4469 on: August 05, 2019, 05:16:43 pm »
Indeed. That is my biggest expectation for the MSO5000 is something proportionately for nothing :)
I wish they’d hurry up and fix the bugs
The following users thanked this post: vk6zgo
Offline TurboTom
« Reply #4470 on: August 05, 2019, 05:40:37 pm »
Exactly!  >:D
P.S. Surely it's the same hardware, that has been sorted a long time ago.
« Last Edit: August 05, 2019, 05:42:09 pm by TurboTom »
Offline Mr. Scram
  • Display aficionado
« Reply #4471 on: August 05, 2019, 06:21:09 pm »
The MSO5000 is frustrating. It's so close to absolutely crushing anything out there yet the slow UI and unsophisticated case design are letting it down. I have the feeling that the difference isn't even very significant in terms of NRE and manufacturing.
« Reply #4472 on: August 05, 2019, 06:40:37 pm »
That was exactly the opinion I had on the thing when I played with it as well.
Offline David Hess
  • DavidH
« Reply #4473 on: August 06, 2019, 06:55:02 am »
David -
So it seems something else is going on although I do not consider that an improvement but rather a mystery.
Dave's reverse engineered schematics show some things which may be related to your test results but I have my doubts that the schematics are completely accurate; they show some odd things.
At some point maybe I can get ahold of one and test it myself.
Offline t_i_t_o
  • Country: bg
« Reply #4474 on: August 06, 2019, 07:50:22 am »
Just to put some oil in the fire, here is the same signal, same Tek probe, but seen with 1.5GHz 8GS/s HP infinium. So compared to the rigol, we see a peak at the beginning of each pulse, where this peak is observed i.e. after the front. On rigol is at the end of the pulse - before the front. Any explanation? Bad interpolation?