Piton:
rob040, when you go to a Chinese site, you need to write not a model, but a series (I have a DS1000Z-E) in the search, which is located at the top - on the right, and then a page with files and firmware only for this series will appear, and all this can be downloaded. If you try to download directly from the main page, where all the series are shown, then registration will be required. I discovered this by accident.
bffargo:
In the RAR file linked by bulba99, there is a .doc file that contains a lot more information than the shorter history notes. It looks to be all internal bug details addressed in each build not the simpler summaries.
Running it through google translate for some of the text that wasn't dual-language already, got the below doc for the changes since Version:00.04.03.02.02 Date:2015-09-07. If you want to see older details just browse the rest of the doc.
It is very weird though the dates don't match up with those in the normal history notes and while some versions exactly match the list of fixes (example: v00.04.04.01.01 has the same details in both but different dates: 2016-08-19 vs. 2016/09/14) others seem totally different. I'm curious why there is such a mismatch?
--- Code: ---Version: 00.04.05.02.02 Date: 2021-09-22
1. Solve the problem of abnormal display when measuring overshoot in some cases
2. The change description table of the new version is at the front.
Version: 00.04.05.02.01 Date: 2021-08-30
1. Add SCPI command to query all product information and determine whether to enter engineering mode
2. Add SCPI command to read self-calibration and low-frequency compensation status
Version: 00.04.05.02.00 Date: 2021-07-24
1. Compatible with Lx25, Lx45 logic chip
2. Open deep storage, advanced triggering, decoding, waveform recording option
Version: 00.04.05.01.00 Date: 2021-06-01
1. Solve the problem that the interface display function is unavailable after default
Version:00.04.05.00.00 Date:2020-12-29
1. Solve the problem of loading the old version ref file, causing the device to hang up M
2. Modify the tracking cursor. When a is out of the screen, modify the value of B and affect the value of A M
3. In the roll mode, the tracking cursor does not move with the waveform after stop M
4. SCPI sends the amplitude parameter of DG module 4.020vpp, and the return value is 4.019 M
5. Revision: FUNC:WREP:FMAX? Command is used to query the current maximum number of playable frames M
6. Modify the problem that trigger Hold off is not applied when the trigger type without triggering Hold off function is switched back to the trigger type with triggering Hold off function M
7. :CURSor:MANual:SOURce set and query command add off parameter to set the source to none M
8. Modify the average mode, Pass fail test channel confusion problem M
9. Modify the maximum gear limit of exp operation to 1gv M
10. When math is closed, math's automatic gear is opened, vertical gear is modified, and automatic gear is updated M
11. Modify the problem of rising edge delay measurement and cursor position deviation in automatic cursor mode M
12. Modify the error of the maximum time measurement value in the automatic cursor mode M
13. Modify the single trigger mode, the last value restart, no waveform trigger problem M
14. Modify the problem of amplitude error when saving the CSV file with channel offset M
15. Modify SPI timeout trigger, when the data bit is less than 8, all bit function settings are unsuccessful M
16. Limit the upper and lower limit of trigger level to ± 5 * vertical gear - vertical offset M
17. Modify XY mode, the full screen and half screen frequently switch, leading to the problem of splash screen M
18. Revision: WAV:XOR? Return value error M
19. Revision: WAV:PRE? Partial return value problem M
20. Modify the problem that the gear is still adjustable when the channel is closed M
21. Modify the welcome page, and the web identification indicator button will not jump M
22. Modify after opening LXI, modify the host name, and the instrument crashes M
23. Fix GPIB communication failure M
24. Add: STOR:IMAGe [:STARt] <file_ Spec > command (same as ds2000), take a screenshot, and save the picture to the path specified in the parameter E
25. Add different edge delay measurement function between channels E
26. Fix LXI socket communication hang up problem M
27. GD spiflash and nandflash are compatible M
28. Factory boot compatible with Gd spiflash and nandflash M
Version:00.04.04.03.02 Date:2017-2-6
1. Fixed the period measure value is error while system is average acquire mode and 12K pts memdepth; M
2. Fixed the phase linear offset of the Filter operation results; M
3. Fixed the error of Math waveform,while system is average acquire mode and horizontal time bass is 500ms ; M
4. Fixed the location error of trigger level,when pressing the Auto button; M
5. Fixed the waveform is error while horizontal time bass is 100ms and 12M pts memdepth,when display in the “X-Y” horizontal mode; M
6. Fixed the overlap error of memory waveform,when the Scope read memory waveform data; M
7. Fixed the problem of save memory data was wrong when the channel invert was opened; M
8. Fixed the event table can’t close while it is openning and Loading waveform file M
9. Fixed the not measured parameters are still displaying ,when the font size after modified M
10. Fixed the return value parameter of XOR can’t refresh correctly whlie Changed the horizontal offset M
11. Fixed the problem of the reference wave of the digital channels position is error; M
12. Fixed the frequency response curve of Digital filter is error; M
13. Fixed the measure error of average parameter when the measured signal is a DC signal; M
Version:00.04.04.03.00 Date:2016-12-20
1. Changed the Prompting English information of LA channel, when closed the button of D7–D0; M
2. Fixed the rules of Pass/Fail created again,when the main time base changed; M
3. Fixed the problem of the position of big grids of main time base is error ,when split display in the “X-Y” horizontal mode ; M
4. Modified a low position of trigger label to the all-measurement window; M
5. Fixed the offset error of memory waveform,when the Scope read memory waveform data; M
6. Fixed the bug of not downloading waveform to signal source; M
7. Fixed the problem of save memory data was wrong when the channel invert was opened; M
8. Fixed the problem of the reference wave of the digital channels position is error while save the reference wave of FFT channel; M
9. Fixed the problem of power off, when horizontal delay is on and opening the waveform recorded ; M
10. Fixed the problem of unable to waveform recording,when main time bass is 100ms and Zoom time base is 100ms; M
11. Translated some traditional Chinese and Korean menus ; M
Version:00.04.04.02.00 Date:2016-10-21
1. The effective value measurement result is disturbed M
2. Self-calibration modification and perfect M
3. Improved LXI, affecting web pages, LAN communication and other related issues M
4. Turn off the low-version boot machine after upgrading, increase the CPU clock M in the APP
Version:00.04.04.01.01 Date:2016-08-23
1. Fixed the Math status of shelter from Measure M
Version:00.04.04.01.00 Date:2016-08-19
1. Supperted the multi-interface of LXI M
2. Fixed the bugs about measure M
Version:00.04.04.00.05 Date:2016-05-30
1. Fixed the bug of not stopping the DSO when Pass/Fail checked failure M
Version:00.04.04.00.04 Date:2016-05-24
1. The menu of decoder can’t refresh correctly when changing the source M
Version:00.04.04.00.03 Date:2016-05-06
1. Fixed the problem of wave and the label of position is error when full display in the “X-Y” horizontal mode; M
2. Fixed the problem of the number of grids of main timebase is error when split display in the “X-Y” horizontal mode; M
3. Fixed the problem of the first frame of record data is error when horizontal delay is on; M
4. Fixed the problem of the label of logic channel is not show when horizontal delay is on; M
5. Fixed the problem of the range of manual cursor is not updated when horizontal delay is on; M
6. Fixed the problem of the line of trigger level disappeared when moved the trigger level to the bottom of screen and horizontal delay was on; M
7. Fixed the problem of the warning tone sounded randomly while the Pass/Fail function wss failure; M
Version:00.04.04.00.00 Date:2016-04-14
1. Increase the function of split display wave in the “X-Y” horizontal mode; E
2. Increase the function of which can use the file name that the user last used in storage module; E
3. Fixed the period measure value is error while system is average acquire mode and 12K pts memdepth; M
4. Fixed the problem of the Math measure items unit is error; M
5. Fixed the problem of the system state is error while undo the “Auto” function; M
6. Fixed the problem of the data type of RS232 decoder event table is error; M
7. Fixed the problem of the system trigger state is error while load the setup file of saved in Single mode; M
8. Fixed the problem of the horizontal offset is error while undo the “Auto” function in system setup of timescale greater or equal to 200ms and horizontal offset greater than 1s; M
9. Fixed the problem of the reference wave of the analog channels position is error while save the reference wave of FFT channel; M
10. Fixed the problem of the slope trigger low level can not be adjusted while the input signal has offset component; M
11. Fixed the problem of the low level of the pulse trigger can not be seted while use the remote command; M
12. Fixed the problem of the setup file can not be loaded by PC software; M
13. Fixed the problem of the decoder can be opened by used the remoted command while the system has not decoder option; M
14. Fixed the problem of the measure item of the pulse counter was wrong; M
15. Fixed the problem of draw cursor position was delayed when adjust the cursor position; M
16. Fixed the problem of save memory data was wrong when the channel invert was opened; M
17. Fixed the problem of read the wave data of memory was wrong when channel 1 and chanel 2 was opened in “Single” mode; M
Version:00.04.03.02.03 Date:2015-09-11
18. Fixed the bug of the trigger option be installed while only the decoder option be installed. M
Version:00.04.03.02.02 Date:2015-09-07
1. Fixed the bug of the REF source can not select the open channels. M
2. Fixed the bug of the function of storage can not store the waveform while analog channel and logic channel are open and the memory depth is 12Mpts. M
3. Fixed the bug of the system can not use the function of Single while the system stoped and trigged. M
--- End code ---
Bicurico:
I would suggest that the mismatch in dates is due to different programmers finishing the fixes of their functions/routines of the FW on different dates.
The FW itself is compiled at a given date, so not always everything adds up.
Also, a FW release fork might be recompiled using the same version main source code, but with some updated libraries.
A modern oscilloscope or spectrum analyzer is nowadays more than just one dedicated machine running a sequentially programmed software. Instead, it is in first place a computer running an operating system, composed of many generic libraries and then there is the specific software with it's own set of libraries. A FW update may contain updates to many different files.
I upgraded the FW of my DS1054Z yesterday with this lates version and it seems to behave OK without issues. I would assume it is perfectly safe to flash the FW.
I have no idea why western Rigol sites do not offer this latest version.
Regards,
Vitor
RoGeorge:
Stupid request, can anybody post a DS1054 print screen with the new blue logo please?
Is that the color for the boot screen Rigol logo, or for the upper left Rigol logo seen on all screens?
I want to see how the new Rigol blue logo looks like. I'll prefer it to be yellow, without messing with firmware customization at each new update. In fact, anything else but blue or magenta would be a OK, IMO.
Adrian_Arg.:
Bicurico said "I updated the FW of my DS1054Z yesterday with this latest version and it seems to behave fine with no problems. I would assume that it is perfectly safe to flash the FW." Bicurico, did you test if the synchronization problem was corrected with captures at 24 megabits?