Author Topic: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO...)  (Read 10847 times)

0 Members and 1 Guest are viewing this topic.

Offline dpenev

  • Regular Contributor
  • *
  • Posts: 195
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO800...)
« Reply #75 on: August 16, 2024, 01:09:13 pm »
Thanks Furna,
I have  followed the procedure I have described two messages back and the log attached.
 

...
sr: scpi_tcp: Successfully sent SCPI command: ':TDIV? '.
sr: scpi: Got response: 'DAT2,#9000000000 AVEDESC1.00E+075.00E-04', length 40.
sr: session: bus: Received SR_DF_HEADER packet.
sr: scpi_tcp: Successfully sent SCPI command: 'WAV:MAXP? '.
sr: scpi: Got response: '5.00E-04', length 8.
sr: scpi: get_int: non-integer rational=500/1000000
sr: session: Could not start siglent-sds device tcp-raw/192.168.1.6/5025 acquisition.
...

'DAT2,#9000000000 AVEDESC1.00E+075.00E-04' looks a concatenation of few responses.

5.00E-04 is my current timing resolution 500us/div

WAV:MAXP?  if sent separately to the oscilloscope returns 10000000 but I don't see it in the log.

 
« Last Edit: August 16, 2024, 01:31:29 pm by dpenev »
 

Offline Furna

  • Regular Contributor
  • *
  • Posts: 146
  • Country: de
  • I still don't know
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO800...)
« Reply #76 on: August 16, 2024, 04:11:12 pm »
Thanks Furna,
I have  followed the procedure I have described two messages back and the log attached.
 

...
sr: scpi_tcp: Successfully sent SCPI command: ':TDIV? '.
sr: scpi: Got response: 'DAT2,#9000000000 AVEDESC1.00E+075.00E-04', length 40.
sr: session: bus: Received SR_DF_HEADER packet.
sr: scpi_tcp: Successfully sent SCPI command: 'WAV:MAXP? '.
sr: scpi: Got response: '5.00E-04', length 8.
sr: scpi: get_int: non-integer rational=500/1000000
sr: session: Could not start siglent-sds device tcp-raw/192.168.1.6/5025 acquisition.
...

'DAT2,#9000000000 AVEDESC1.00E+075.00E-04' looks a concatenation of few responses.

5.00E-04 is my current timing resolution 500us/div

WAV:MAXP?  if sent separately to the oscilloscope returns 10000000 but I don't see it in the log.

Can you try with a different resolution? I was succesfull with 10mV/div with SDS824X HD

See also
https://www.eevblog.com/forum/testgear/new-sigrokpulseview-hardware-support-(siglent-sds-hd-rigol-dho800-)/msg5605601/#msg5605601
With SDS2104X+ it is not possible to go lower than 10mV/div and fredo_ is looking at it.
« Last Edit: August 16, 2024, 04:12:57 pm by Furna »
The sky above the port was the color of television, tuned to a dead channel.
 
The following users thanked this post: fredo_

Offline dpenev

  • Regular Contributor
  • *
  • Posts: 195
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO800...)
« Reply #77 on: August 17, 2024, 12:40:08 pm »
Seems not related with Voltage resolution. I have tested with 1V/div and still similar behavior.
| am attaching another log for someone to check. I have a feeling that it is a timing issue in the the communication.
Assuming that the software is working fine for someone else with SDS2000X+
In my case both the PC running PV and the scope are in the same LAN connected directly in the same router, 1,2m LAN cables.

Pinging 192.168.1.6 with 32 bytes of data:
Reply from 192.168.1.6: bytes=32 time<1ms TTL=64
Reply from 192.168.1.6: bytes=32 time<1ms TTL=64
Reply from 192.168.1.6: bytes=32 time<1ms TTL=64
Reply from 192.168.1.6: bytes=32 time<1ms TTL=64
 
The following users thanked this post: fredo_

Offline fredo_Topic starter

  • Regular Contributor
  • *
  • Posts: 86
  • Country: fr
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO800...)
« Reply #78 on: August 17, 2024, 12:57:16 pm »
Hi @dpenev and thank you for your investigations !

Seems not related with Voltage resolution. I have tested with 1V/div and still similar behavior.
| am attaching another log for someone to check. I have a feeling that it is a timing issue in the the communication.
Assuming that the software is working fine for someone else with SDS2000X+
In my case both the PC running PV and the scope are in the same LAN connected directly in the same router, 1,2m LAN cables.
No it doesn't seem to be a timing issue, the problem comes from the response of the scope to the ACQ:POIN? command.
You scope gives a ".9" reply while the Sigrok (and the documentation) expects something of the form "1.25E+08".
This command returns the number of points of the current waveform. Can you check the memory depth of the scope and maybe the value returned by ACQ:POIN? SCPI command form web control ?
« Last Edit: August 17, 2024, 01:14:04 pm by fredo_ »
 
The following users thanked this post: Furna

Offline fredo_Topic starter

  • Regular Contributor
  • *
  • Posts: 86
  • Country: fr
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO800...)
« Reply #79 on: August 17, 2024, 10:34:10 pm »
Oh and @dpenev, you may want to try and reboot your scope (including removing the powder cable).
It happened to me a few times: I started having very strange SCPI responses from the scope, and after a full reboot, everything went back to normal  :-//.
« Last Edit: August 18, 2024, 06:14:19 am by fredo_ »
 

Offline fredo_Topic starter

  • Regular Contributor
  • *
  • Posts: 86
  • Country: fr
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO800...)
« Reply #80 on: August 26, 2024, 09:20:58 pm »
I couldn't set the vertical sensitivity lower than 10mV/div.
Both scope can go down to 500µV/div so it seem that PulseView
doesn't support it even it is displayed in configure channel top menu.
That I should be able to fix, I'll take a look.
That would be great !

Hi there,
Here is a new version that should fix the sub 10mV vertical range issue for Siglent models : Windows (64bits) installer (zip password is 'sigrok').

08/26/2024 Update
Siglent SDS : fixed support for sub 10mV vertical ranges

 
The following users thanked this post: KungFuJosh, Furna, qubit

Offline Spike

  • Regular Contributor
  • *
  • Posts: 77
  • Country: us
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO800...)
« Reply #81 on: August 26, 2024, 10:49:07 pm »
I couldn't set the vertical sensitivity lower than 10mV/div.
Both scope can go down to 500µV/div so it seem that PulseView
doesn't support it even it is displayed in configure channel top menu.
That I should be able to fix, I'll take a look.
That would be great !

Hi there,
Here is a new version that should fix the sub 10mV vertical range issue for Siglent models : Windows (64bits) installer (zip password is 'sigrok').

08/26/2024 Update
Siglent SDS : fixed support for sub 10mV vertical ranges

Updated macOS builds available; see first post for link.
 
The following users thanked this post: kripton2035, fredo_

Offline hansibull

  • Regular Contributor
  • *
  • Posts: 125
  • Country: no
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO...)
« Reply #82 on: August 30, 2024, 07:50:23 am »
Amazing work! What are the odds of getting your fork merged into the official Sigrok repo? I have been using Sigrok's libserialport library, and I even found a nasty bug that affects macOS. I clearly demonstrated the issue and even submitted a PR, but It hasn't gotten any attention from the developers.
 
The following users thanked this post: fredo_

Offline fredo_Topic starter

  • Regular Contributor
  • *
  • Posts: 86
  • Country: fr
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO...)
« Reply #83 on: August 30, 2024, 09:05:45 am »
Amazing work! What are the odds of getting your fork merged into the official Sigrok repo?
Thanks !
Well I created the two PRs yesterday, now that things seem quite stable:
https://github.com/sigrokproject/libsigrok/pull/245
https://github.com/sigrokproject/libsigrok/pull/246
So wait and see !...
 
The following users thanked this post: thm_w, qubit

Offline Furna

  • Regular Contributor
  • *
  • Posts: 146
  • Country: de
  • I still don't know
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO800...)
« Reply #84 on: August 30, 2024, 11:09:43 am »
Regarding LA acquisition, I spent quite some time trying and understand how to have pagination work (data blocks are limited to 5Mpts, over that memory depth, data has to be paginated) but I came to the conclusion that the SDS2000X HD has a bug that prevents pagination to work. The result is that the first 5Mpts datablock is replicated several times on LA channels :-/
I ended up sending a mail to Siglent support to report the issue. Not sure if/when I'll get a response and this will be fixed...
Just a quick update on this , I got the following  reply from Siglent support :
Quote
This may be a bug in the software. Our R&D department is analyzing this problem. I will inform you as soon as I have the result.
Wait and see !...

Any update from Siglent ?
The sky above the port was the color of television, tuned to a dead channel.
 
The following users thanked this post: fredo_

Offline Furna

  • Regular Contributor
  • *
  • Posts: 146
  • Country: de
  • I still don't know
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO...)
« Reply #85 on: August 30, 2024, 11:10:28 am »
Amazing work! What are the odds of getting your fork merged into the official Sigrok repo?
Thanks !
Well I created the two PRs yesterday, now that things seem quite stable:
https://github.com/sigrokproject/libsigrok/pull/245
https://github.com/sigrokproject/libsigrok/pull/246
So wait and see !...

Let us know if we can sponsor your PR
The sky above the port was the color of television, tuned to a dead channel.
 
The following users thanked this post: fredo_

Offline fredo_Topic starter

  • Regular Contributor
  • *
  • Posts: 86
  • Country: fr
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO...)
« Reply #86 on: August 30, 2024, 11:53:00 am »
Any update from Siglent ?
Yes I had a first reply from the support team with a code snippet showing a working example of pagination on digital data, but this was not relevant since it was paginating only on a few hundred of bytes.
I replied explaining that the problem only appeared for 5Mpts+ frame, but they did not come back to me so far (this was mid-august).

Let us know if we can sponsor your PR
For now I'm trying to reach out to project committers on the dedicated Discord channel...

 
The following users thanked this post: Furna

Offline hansibull

  • Regular Contributor
  • *
  • Posts: 125
  • Country: no
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO...)
« Reply #87 on: August 30, 2024, 01:25:09 pm »
I really hope your PRs gain some traction from the Sigrok maintainers. They cover the most popular Rigol and Siglent scopes currently on the market, making Sigrok relevant for many new users.
I used your Sigrok fork to capture data with a Siglent SDS2000X HD and decode an SPDIF signal, something no scope has built-in decoding for.
 
The following users thanked this post: Furna, fredo_

Offline Spike

  • Regular Contributor
  • *
  • Posts: 77
  • Country: us
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO...)
« Reply #88 on: August 30, 2024, 05:34:07 pm »
Amazing work! What are the odds of getting your fork merged into the official Sigrok repo?
Thanks !
Well I created the two PRs yesterday, now that things seem quite stable:
https://github.com/sigrokproject/libsigrok/pull/245
https://github.com/sigrokproject/libsigrok/pull/246
So wait and see !...
The reviewer of the siglent PR on which your work is based was github user fenugrec.  I’d be shocked if that isn’t the very same fenugrec user as here on the EEVblog forum.  Perhaps you could reach out here and see if you can get things moving a little sooner.
 
The following users thanked this post: fredo_

Offline wiretail

  • Newbie
  • Posts: 2
  • Country: us
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO...)
« Reply #89 on: September 21, 2024, 09:40:55 am »
Just wanted to add a hearty thanks for this.  One of the first things I tested with my new SDS804/824.  Works great!

Doing analog->digital convert->custom python decoder's direct from the scope, , is wonderful.  (Wherever it might be -- with a tailscale network, it's handy for remote investigation too.)
 
The following users thanked this post: fredo_

Offline fredo_Topic starter

  • Regular Contributor
  • *
  • Posts: 86
  • Country: fr
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO...)
« Reply #90 on: September 21, 2024, 09:48:05 am »
Just wanted to add a hearty thanks for this.  One of the first things I tested with my new SDS804/824.  Works great!

Doing analog->digital convert->custom python decoder's direct from the scope, , is wonderful.  (Wherever it might be -- with a tailscale network, it's handy for remote investigation too.)
Thanks for the feedback @wiretail !
If you have some spare time, you might want to check the other project I'm working on: https://www.eevblog.com/forum/testgear/ngscopeclient-ultimate-acquisition-software-protocol-decoder-debug-tool/
We need testing from SDS800 owners ;-)
 
The following users thanked this post: Furna

Offline fredo_Topic starter

  • Regular Contributor
  • *
  • Posts: 86
  • Country: fr
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO800...)
« Reply #91 on: September 25, 2024, 09:22:01 pm »
Regarding LA acquisition, I spent quite some time trying and understand how to have pagination work (data blocks are limited to 5Mpts, over that memory depth, data has to be paginated) but I came to the conclusion that the SDS2000X HD has a bug that prevents pagination to work. The result is that the first 5Mpts datablock is replicated several times on LA channels :-/
I ended up sending a mail to Siglent support to report the issue. Not sure if/when I'll get a response and this will be fixed...
Just a quick update on this , I got the following  reply from Siglent support :
Quote
This may be a bug in the software. Our R&D department is analyzing this problem. I will inform you as soon as I have the result.
Wait and see !...
Any update from Siglent ?
I just received a confirmation from Siglent support that they could reproduce the bug and that they are working on a fix !  :-+
 
The following users thanked this post: kripton2035, Furna

Offline Furna

  • Regular Contributor
  • *
  • Posts: 146
  • Country: de
  • I still don't know
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO800...)
« Reply #92 on: September 25, 2024, 09:24:25 pm »
I just received a confirmation from Siglent support that they could reproduce the bug and that they are working on a fix !  :-+

Thank you for the update!
The sky above the port was the color of television, tuned to a dead channel.
 
The following users thanked this post: fredo_

Offline qubit

  • Newbie
  • Posts: 2
  • Country: de
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO...)
« Reply #93 on: October 13, 2024, 05:07:33 am »
I've just been scrolling through the open pull requests on github and there is actually someone working on fixing the Siglent SDS E-series driver, awesome!
Here's the link to the PR: https://github.com/sigrokproject/libsigrok/pull/247
Anyone with an E/U-series oscilloscope can help with testing the driver, and anyone with a 1000X/2000X series scope can help with regression testing  :)
@fredo_ if you have time, could you please kindly share a compiled version of the PR247 so that we can test the changes? I'm still not able to get the building process working  :-\
« Last Edit: October 13, 2024, 05:16:44 am by qubit »
 

Offline fredo_Topic starter

  • Regular Contributor
  • *
  • Posts: 86
  • Country: fr
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO...)
« Reply #94 on: October 14, 2024, 10:25:15 am »
I've just been scrolling through the open pull requests on github and there is actually someone working on fixing the Siglent SDS E-series driver, awesome!
@fredo_ if you have time, could you please kindly share a compiled version of the PR247 so that we can test the changes? I'm still not able to get the building process working  :-\
Hi @qubit,

Yes I'm aware of this PR and already proposed to merge it with mine on Sigrok Discord channel.
They are preparing a release and plan to integrate some PRs to it, Rigol and Siglent driver update PRs (245,246,247) are candidate for it: https://sigrok.org/wiki/Developers/Release_0.8.0
TBH I'm not spending more time on it until I'm sure this eventually make its way to upstream ; I moved to another project (ngscopeclient) for now.
I can try and guide you to have your build work thought, what platform are you using?
 

Offline DirtySignal

  • Newbie
  • Posts: 4
  • Country: za
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO...)
« Reply #95 on: October 14, 2024, 01:41:46 pm »
I've just been scrolling through the open pull requests on github and there is actually someone working on fixing the Siglent SDS E-series driver, awesome!
Here's the link to the PR: https://github.com/sigrokproject/libsigrok/pull/247
Anyone with an E/U-series oscilloscope can help with testing the driver, and anyone with a 1000X/2000X series scope can help with regression testing  :)
@fredo_ if you have time, could you please kindly share a compiled version of the PR247 so that we can test the changes? I'm still not able to get the building process working  :-\

I have a DSD1104X-E... do you have a link for a test binary on Windows? Thanks
 

Offline fredo_Topic starter

  • Regular Contributor
  • *
  • Posts: 86
  • Country: fr
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO...)
« Reply #96 on: October 14, 2024, 01:46:31 pm »
I have a DSD1104X-E... do you have a link for a test binary on Windows? Thanks
Yes you'll find a link in the first page of this topic : https://www.eevblog.com/forum/testgear/new-sigrokpulseview-hardware-support-(siglent-sds-hd-rigol-dho800-)/
Please report the results of your tests here as it will help validate the corresponding Pull Request and merge the code to Sigrock upstream.
 

Offline DirtySignal

  • Newbie
  • Posts: 4
  • Country: za
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO...)
« Reply #97 on: October 14, 2024, 01:49:23 pm »
Yes you'll find a link in the first page of this topic : https://www.eevblog.com/forum/testgear/new-sigrokpulseview-hardware-support-(siglent-sds-hd-rigol-dho800-)/
Please report the results of your tests here as it will help validate the corresponding Pull Request and merge the code to Sigrock upstream.

I think I was not clear enough, apologies - I meant a test binary for the PR mentioned in the previous post. I've tested the binary you provided in the first post but sigrok hangs when trying to acquire anything from the scope... so I wanted to test that PR which seems to be a fix for that issue specifically.

I had a look at building from source but it looks a little complicated for a background activity at the moment...
 

Offline fredo_Topic starter

  • Regular Contributor
  • *
  • Posts: 86
  • Country: fr
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO...)
« Reply #98 on: October 14, 2024, 02:45:41 pm »
Oh no sorry it's me who misread your model number, I thought it was an old Rigol.
I don't have a binary for PR #247, sorry, maybe you should get in touch with its author for that ?: https://github.com/mike42
 
The following users thanked this post: DirtySignal

Offline DirtySignal

  • Newbie
  • Posts: 4
  • Country: za
Re: New Sigrok/PulseView hardware support (Siglent SDS HD, Rigol DHO...)
« Reply #99 on: October 14, 2024, 02:52:26 pm »
Oh no sorry it's me who misread your model number, I thought it was an old Rigol.
I don't have a binary for PR #247, sorry, maybe you should get in touch with its author for that ?: https://github.com/mike42

Thanks, I fired up a WSL session busy building the dev environment, let's see how far I get :)

edit: Windows 32bit and 64bit binary built successfully for that PR
« Last Edit: October 15, 2024, 01:47:22 pm by DirtySignal »
 
The following users thanked this post: fredo_


Share me