| Mahagam:
Nothing common with SPD5000 series commands. For SPD1168X you can set/get voltage/current with fully formatted string e.g.: ch1:volt 5.000 ch1:volt? 5.000 ch1:curr 0.5 ch1:curr? 0.5 but you cannot set/get ocp in this way, no ch1:ocp 2.345 will be accepted. But!!! you can just set/get voltage/current/ocp via simplest format! : volt 4.534 volt? 4.534 curr 0.6 curr? 0.6 ocp 2.0 ocp? 2.0 *idn? Siglent Technologies,SPD1168X,SPD13DCD7R1985,2.1.2.11,V2.0 As you can see, I have SPD1168X with HW version 2.0 and fw version 2.1.2.11 |
| Mahagam:
I still have a question about SPD1168X. I have siglent 4 devices (SDS2104X+, SPD1168X, SDG2042X, SDM3055). Everything is connected to a LAN, but only 3 devices can be detected via VXI-11. For the SPD1168X I have to enter IP address manually. Is there any way to find SPD online? |