Nothing common with SPD5000 series commands.
For SPD1168X you can set/get voltage/current with fully formatted string e.g.:
ch1:volt 5.000
ch1:volt?
5.000
ch1:curr 0.5
ch1:curr?
0.5
but you cannot set/get ocp in this way, no ch1:ocp 2.345 will be accepted.
But!!!
you can just set/get voltage/current/ocp via simplest format! :
volt 4.534
volt?
4.534
curr 0.6
curr?
0.6
ocp 2.0
ocp?
2.0
*idn?
Siglent Technologies,SPD1168X,SPD13DCD7R1985,2.1.2.11,V2.0
As you can see, I have SPD1168X with HW version 2.0 and fw version 2.1.2.11