Products > Test Equipment
New UNI-T UT892 2000V AC/DC High Voltage Multimeter
PartialDischarge:
So here are some more tests
For the 1000V input, AC test up to 2kVrms, there is a limiting device that trips at around 1.9kVrms:
For the 2000V input, AC test up to 4kVrms, no problems, although this is with the 760k resistor up front instead of the original PTC:
For the 2000V input, AC test up to 5kVrms, no problem, current meter reads 250uA as expected, I don't really see the point in going further, 5kVrms is 7kV peak.
joeqsmith:
Thanks for taking the time to make these clips. Would you consider getting a second meter to repeat the tests without the mods? May be interesting to see you sweep the voltage as well.
PartialDischarge:
--- Quote from: joeqsmith on September 10, 2022, 02:43:22 pm ---Thanks for taking the time to make these clips. Would you consider getting a second meter to repeat the tests without the mods? May be interesting to see you sweep the voltage as well.
--- End quote ---
I don't just need another one cause this one is fine and I'm going to use it, but I can short the PTC and test again, that'll be easier.
joeqsmith:
No concerns about the unknown resistors?
PartialDischarge:
--- Quote from: joeqsmith on September 10, 2022, 05:25:12 pm ---No concerns about the unknown resistors?
--- End quote ---
I'm more concerned about a 2 inch trace running in the PCB from the input. Resistors if they are bad they will change their properties a lot but they are not going to suddenly short to 0.
Navigation
[0] Message Index
[#] Next page
[*] Previous page
Go to full version