A device rated at 50MHz should be usable at that frequency, no? Even of the fast mode this struggles beyond 47~MHz and 50 is outright unusable.
There are lots of weird trigger bugs and just general trigger issues. Trigger missing the signal instead of the rising edge being centered. CH1 not centering even when CH2 is off.
The amplitude of signals are sometimes completely off between triggers (halved, or worse)? Then they're correct again on another random trigger.
Firmware v1.64, apparently the latest.
Do I have a lemon or is this the general experience? I'm leaning towards initiating a return while I can because for actual use, this is worthless at the current state. It's a toy.
No issues with the DMM but this isn't worth it as an 80$ DMM with a broken OSC.
Agree with all you are writing - however I find v1.4.2 the best version, I have tried.
The oscilloscope IS usable, I think, but you have to know its limitations for now - and hope ZOYI are reading all what comes up in this forum for releasing a much better "final" FW.
It certainly have its trigger issues, but the way I work with it is to avoid triggering below 500nsec - instead I trigger at 500nsec, "Hold" the signal and zoom in to 250-10nsec and "Move" the signal to what I want to look at.
From 25usec timebase and all the way down to 10nsec, the scope will sample at 5nsec intervals (200MHz sample rate), so no data are lost by triggering at 500nsec - and the full sample window is about 600usec in total from 25usec to 10nsec.
The best part is the portability and battery powered and free from relating to ground. But you can also get this from other small scopes
If you come across a scope that is below 100€ and better performance and more "final" firmware, it would be really interesting to hear about