The measurement values in Scope mode seem to have more problems than I thought before when in DC input mode(FW1.32). I measured the switching on signal of my PSU at 12V setpoint. I set the scope to 100ms/div and 5V/div. The signal on the screen clearly shows the 12V amplitude of the voltage ramp. The vpp,max,min values are completely off and do depend on where I set the vertical offset.
Zoyi please fix this. I cannot imagine that this is hard to do.

looks like this issue is fixed in fw 1.35. The min/max measurements still give funny results in roll mode and stopped. Btw, the trigger level value shown at the top is nonsense. Note , the trigger vert position on the Zoyi is always related to the screen coordinates. When you switch the vert sensitivity , the trigger arrow stays at the same position. A bit unusual for me, but for this small unit maybe not a bad choice. 
« Last Edit: April 26, 2024, 02:54:30 pm by ksxx »
 

Offline ksxx

  • Contributor
  • Posts: 30
  • Country: de
I did some tests with the Zoyi and some other meters I have access to. The AD780 was trimmed to give a 2.5000V output on the (old) HP desktop meter. The film capacitors were measured with an Agilent 4294A impedance analyser at 10KHz. The table shows that my 703S gives the results closest to the ones by the more advanced equipment. Yet most of the meters tested do well enough for most situations. Even the 10$ T7 component tester gives great results!

 
The following users thanked this post: black6host, Razvan_N

Offline lazalem

  • Newbie
  • Posts: 1
  • Country: ph
Good Day!

Just want to know if ZT-703S is a good multimeter+oscilloscope for a hobbyist like me.  I am coming from a DSO150.  Which is a better buy among these:

Zoyi ZT-703S ~ 37.80USD (New)
Owon HDS242S ~ 74.16USD (New)
Owon HDS2202S ~ 78.07USD (Used)

At double the price, does the owon variants have better features, accuracy, reliability?


Thanks!
 

Offline Vapor8

  • Newbie
  • Posts: 5
  • Country: us
I did some tests with the Zoyi and some other meters I have access to. The AD780 was trimmed to give a 2.5000V output on the (old) HP desktop meter. The film capacitors were measured with an Agilent 4294A impedance analyser at 10KHz. The table shows that my 703S gives the results closest to the ones by the more advanced equipment. Yet most of the meters tested do well enough for most situations. Even the 10$ T7 component tester gives great results!

Would you mind sharing what version firmware the Zoyi 703S used in your testing has? Thanks!
 

Offline ksxx

  • Contributor
  • Posts: 30
  • Country: de
Would you mind sharing what version firmware the Zoyi 703S used in your testing has? Thanks!
fw was 1.32. I now tried 1.35 and there is no change in the results (as I would assume).
 
The following users thanked this post: Vapor8, black6host

Offline lchao

  • Contributor
  • Posts: 12
  • Country: at
Good Day!

Just want to know if ZT-703S is a good multimeter+oscilloscope for a hobbyist like me.  I am coming from a DSO150.  Which is a better buy among these:

Zoyi ZT-703S ~ 37.80USD (New)
Owon HDS242S ~ 74.16USD (New)
Owon HDS2202S ~ 78.07USD (Used)

At double the price, does the owon variants have better features, accuracy, reliability?


Thanks!

Where did you find these prices? I can find just these lowest prices:
Zoyi 703s shipped to the Philippines is USD 75 on Aliexpress,
Owon HDS242S is USD 144 on Banggood at the moment
Be careful, because sometimes the price os other models are shown if you search for these products (e.g. Zoyi 702 and Owon HDS242 - which doesn't have signal generator)

The 2 products are close to each other in specs, but Owon is double the price. I'm not an electornics expert at all, but based on what I've read so far I found the Owon HDS242S far better than the Zoyi 703s.

Pros for Owon HDS242S compared to Zoyi 703s (for me):
- It seems it has the same hardare as the 70 MHz model, therefore the signal amplitude doesn't shrink even when measuring above 30MHz
- Usable signal generator (Zoyi 703s is very noisy -> low quality, signal output is not BNC)
- Works as described, it meets the specifications
- Has more probe settings e.g. x1000, x10000  (Zoyi has x100 starting since a recent firmware update)
- 4400 mAh battery capacity - 2x 18650 battery (vs. 3400 mAh, 1x 18650 battery in Zoyi)
- Horizontal scale: 5ns/div-1000s/div (Zoyi has 10ns/div-20s/div)
- Stable and accurate software (Zoyi has a lot of inaccuracy issues in the measurements)

Cons for Owon HDS242S compared to Zoyi 703s:
- Has only 8K memory (Zoyi has 128K)
- Multimeter part has "just" 20 000 counts (Zoyi has 25 000)
- It seems firmware is not being updated (Zoyi is continuously gettig FW updates - unfortunately it doesn't mean that all known issues are fixed)
- Somewhat lower bandwidth based on specs: 40 MHz (Zoyi _claims_ 50 MHz, but while you can measure in Owon's full specified bandwidth, with Zoyi's it's not the case)
- Only sold with 1 probe (Zoyi is sold with 2 probes as an option)

All in all I found the HDS far more superior than the Zoyi, but it also comes with an almost 2x price tag.

DISCLAIMER: I have very limited knowledge in electronics. This is solely my opinion and understanding based on the information from manufacturers or elsewhere I found on the Internet. I don't own any of these products.
 

Online jeremybh1

  • Contributor
  • Posts: 15
  • Country: au
Just as I was warming to this device it's let me down in testing in-circuit AC readings.
It shows a voltage then goes to 0, setting the range manually does nothing.
Use case: Macintosh Plus vintage computer power supply.  Visual map of what expected in-circuit voltages should be in both AC and DC.
Uni-T 61 shows expected 20VAC on cathode and Zoyi 703 just couldn't lock on. 
Was able to repeat this failure of flashing a reading then going to 0VAC on other test points too.

So I can't quite use this as a multimeter
 

Online csuhi17

  • Regular Contributor
  • *
  • Posts: 229
  • Country: hu
Owon is more imprecise, but more reliable in my opinion.

Owon also has hardware errors, the HDS2202 series drains the battery quickly, consumes 1.3A current.
There are times when the message Low battery appears on the display, but the device continues to operate for an hour or so. :scared:
There is sometimes a slight DC offset, perhaps due to heating.


But it is true that the function generator is very good.
Its structure creates a professional feeling, it is quite tough.
It also has some good sides, for example 1Gs/s even with two channels. 8)
But for the rest, sampling is halved! From HDS242 to HDS2102

For 78 USD, I think it's worth it, even if you look at the small memory 8k. This is just my personal opinion.
I have not yet seen a video in which they tested whether the Zoyi use its 128k memory in all time bases.
Fnirsi has already surprised me... In a bad way... :palm:

To be honest, I have not yet seen a study in which they discuss the minimum amount of memory required for different measurements. :-//
 