OK, this is extremely helpful. I hadn't tried a single scan w/trigger. I like it. Thank you.
So, one thing that would be helpful is a display top/left? of the actual time post-trigger that is being displayed as the trace is scrolled back and forth.
When it is scrolled back and forth, it would be great if:
           There was a cursor marking the delta time since the trigger or perhaps the beginning of capture (configurable), and
           Wrapping around the trace as it is scrolled? I think there should be a feature setting to turn that on/off.
Online csuhi17
  • Country: hu
« Reply #601 on: September 11, 2024, 04:44:26 pm »
Normally, the manufacturer should set a limit, Owon does not allow you to scroll any further, if you set the time base higher, you only see the signal in the memory, and next to it an empty field on two sides.
This is not a user error, but a Zoyi bug.
Fnirsi oscilloscope = waste&regret
or maybe I was wrong
The following users thanked this post: xKertx, apollo11fan
Offline apollo11fan
« Reply #602 on: September 11, 2024, 07:52:43 pm »
FWIW - if there is someone looking to get some basic background information on Oscilloscopes, this Tektronix guide is free for downloading
https://download.tek.com/document/03W_8605_7_HR_Letter.pdf
The following users thanked this post: Razvan_N, lchao, Astur_TorQue
Offline xfint34
« Reply #603 on: September 11, 2024, 08:28:46 pm »
The currently selected cursor line disappears when time division is >= 200ms. It makes the cursor useless since I can't see where I'm trying to position it. The bug exists back to at least firmware 1.3.2.
The following users thanked this post: apollo11fan
Offline xKertx
« Reply #604 on: September 20, 2024, 04:43:19 pm »
I wanted to use the Rigol PVP3150 150MHz probe with this scope because it has 20MHz bandwidth in 1X mode and i thought it would be good for low voltages.
And while it seems to have much less noise, for some reason in 10X mode it shows annoying overshoot at the edges and the compensation adjustment wont help much.
The compensation kind of works, but the spike at the beginning remains no matter whitch position i turn the compensation screw.
Any ideas what could cause this?
I would have thought that a probe, that costs almost as much as this oscilloscope, would perform flawlessly. But atleast in this case the cheap probe that came with the scope works better.
Or, does it have something to do with the fact that the more expensive Rigol probe probably has lower capacitance and the cheap probe is somehow matched with the scopes input capacitance?
On the screenshot the CH1 is the probe that came with the ZT-703s and CH2 is the Rigol PVP3150.
« Last Edit: September 20, 2024, 06:46:13 pm by xKertx »
Offline nikbry
« Reply #605 on: September 20, 2024, 05:46:09 pm »
FWIW I just tested with the only other probes I have that came with my Owon VDS1022i USB scope and they work just like the Zotek one. The Owon has a similar bandwidth spec, so I assume the probes are similarly specd.
Online csuhi17
  • Country: hu
« Reply #606 on: September 21, 2024, 06:04:31 am »
Yes, there are some that cannot be compensated with it, I did this with the Fnirsi scope, I could not compensate with the factory one, and I had to sacrifice one of the Owons.
I tried a few that were at hand, and the Owon 350MHz undercompensates, the Rigol PVP2350 is good for me and other noname too.
He also writes the compensation value on his data sheet. for PVP2350 is from 6pF to 24pF.
Offline Wintel
The following users thanked this post: jellytot, gtroc71, csuhi17
Offline bebopdk
« Reply #608 on: September 21, 2024, 08:28:18 am »
Link is dead.
Offline jellytot
  • Country: ie
« Reply #609 on: September 21, 2024, 08:54:01 am »
copy and pasted link in new window worked for me! Initial thoughts are it seems to be improved, triggering, stability, etc..
Offline bebopdk
« Reply #610 on: September 21, 2024, 09:46:26 am »
 :-+Thanks, that did the trick...
Online csuhi17
  • Country: hu
« Reply #611 on: September 21, 2024, 09:54:42 am »
The generator function is stuck for me.
Same with the previous firmware.
OutPut is set to square wave, the three parameters for each ---
When I go to MoreAPPs->GEN, it is active.
I can't turn it off, if I change one of the parameters, it always goes back to sine after i exiting and entering again.
Version V1.0.1.
Has anyone had a similar problem, did you manage to fix it?
Offline nikbry
« Reply #612 on: September 21, 2024, 10:35:49 am »
Version V1.0.1.
Not sure I understand....here's how it works for me:
When I press Menu I have two ways of generating a signal: 1) Output Setting 2) More Apps Enter/GEN. I think they are completely separate. Option 1) let's you use the scope simultaneously, option 2 doesn't.
When I use More apps/GEN, with the up down cursor I choose the waveform, pressing Menu switches the output ON/OFF, and F1 through F4 and left right cursor change the parameters. To exit the GEN mode I need to press the MODE button and the signal generation stops.
I'm on FW 1.32 GEN V 1.0.1
Not sure if this answers your question!?
Online csuhi17
  • Country: hu
« Reply #613 on: September 21, 2024, 11:15:46 am »
Yes, thank you.
I didn't even dare to think that the signal generator part only works separately, I've watched videos and reviews of many cheap scopes, but so far this is the first one where you can't check the signal of the signal generator with its own scope. ???
Apart from the OutPut setting on the scope part, which in turn limits the properties of the scope.
I don't remember that this thing was mentioned in the comments, that is, that the scope and the GEN app do not work at the same time.
Then I'm happy, it's not that my device is defective. :phew:
Offline lchao
  • Country: at
« Reply #614 on: September 21, 2024, 11:36:31 am »
Hi, I've recently put my hand on an XEAST XE703S, which is just another rebranded version of the ZOYI dual channel scopemeter.
Specs, description and EN manual: https://www.xeastek.com/xeast-thermal-imager-for-hunting-xe-c18-c19-day-night-aiming-lens-outdoor-adjustable-focus-clarity-lens-infrared-thermal-camera-1709356654564356.html
It came with the v1.3.8 firmware (APP) and v1.0.16 BOOT.
It was a real bargain for ~50 USD (with 2 probes, incl shipping) as a special offer on TEMU, thanks to Electronics Repair School YT channel .
V1.4.2 FW is also available from here: http://www.szzotek.com/h-col-159.html (stay on the China version and use translator). But I couldn't find release notes.
Offline RAPo
« Reply #615 on: September 21, 2024, 11:47:26 am »
Thanks. The update went smoothly.
Only when problem when no signal on channel 1, pressing auto set the trigger to channel 2 even when that channel is off.
Offline jellytot
  • Country: ie
« Reply #616 on: September 21, 2024, 12:09:59 pm »
Have you set "Trig Source" from the menu ?
Offline RAPo
« Reply #617 on: September 21, 2024, 12:28:06 pm »
yes.
Offline jellytot
  • Country: ie
« Reply #618 on: September 21, 2024, 01:09:04 pm »
yes.
Ok. It's Working fine for me. Maybe a factory reset would help?
Online csuhi17
  • Country: hu
« Reply #619 on: September 21, 2024, 01:31:13 pm »
It's the same for me.
With active CH1 and inactive CH2, the source of the trigger is CH1.
There is no signal on either input and I press AUTO, then the trigger source switches to CH2.
If I press the MENU button and use the arrow to navigate to where I can set the trigger source, it switches back to CH1 automatically.
a little interesting, but not a big deal.
If I noticed correctly, AUTO now places it in the middle like the two channels, and you can finally see the signal in normal size, you don't need to adjust it. :clap:
« Last Edit: September 21, 2024, 05:27:51 pm by csuhi17 »
The following users thanked this post: RAPo
Offline nikbry
« Reply #620 on: September 21, 2024, 02:52:04 pm »
Are you saying Auto Range works well now?
Offline RAPo
« Reply #621 on: September 21, 2024, 03:27:25 pm »
No auto doesn't work well. See the image. And it is slow. You can see on the screen that it is running over the V/div settings very slowly.
Offline indman
  • Country: by
« Reply #622 on: September 21, 2024, 03:38:53 pm »
Slow and clumsy - these are the definitions of oscilloscope operation, which are the most suitable for ZT-703S.
The space of the large display is not utilized at all efficiently. The windows with measurement results only interfere with the normal perception of the signal. In general, after a long enough use of this device as an oscilloscope my opinion - slow, inconvenient, unpleasant. Unlike a multimeter, which is really convenient to use.
Offline nikbry
« Reply #623 on: September 21, 2024, 04:44:02 pm »
Is this with FW 1.42?
Offline nikbry
« Reply #624 on: September 21, 2024, 04:47:01 pm »
What are you comparing it to, to have such a negative opinion?