When I set the current set point to 20mA it immediately went into CC mode without anything connected to the terminals.
Does anyone else have experience with this behavior?
When I got my NGE103 I noticed that behavior on CH1 and I got in contact with the R&S support.
I was told to send it back and just got a replacement unit delivered today. Unfortunately, it has the same issue but this time on CH3.
Video of the issue on my first unit:
https://youtu.be/ltwoPL4OYWsI also noticed that CH2 & CH3 enter CC-mode about 15mA before actually reaching the set current.
I use an active load to apply a constant resistive load to the supply's output.
As you can see in the video, I increase the load until about 85mA where it already shows cc-mode (set current is 100mA). But it looks like it's constantly trying to jump between CV and CC.
Only if I increase the load further, it will eventually limit to the set current and decrease the voltage without jumping around.
Video:
I don't know if this is a hardware issue or software related, I got the unit delivered with FW version 1.56 and tried updating to 1.57, but the same issue still persists.
Not sure what to do now, with this being the 2nd unit that has issues and reading previous post in this thread...
I mean currently I would trust my two Korad supplies over this any day, and they were like 1/10th the price
EDIT:
I turned the supply on today and now I have problemes even with CH1, that I didn't have when I tested yesterday!
Video:
https://youtu.be/CH3KNrPozwsLike I already mentioned above: CH3 stays at 0V, even with nothing connected to the output, until I increase the set current to above 15mA...
But now CH1 won't start limiting the current until around 40mA, with a set current of only 3mA??? It didn't do this yesterday!
Video:
https://youtu.be/NoL4lNmjB5cAnother test with the set current at 100mA: It won't start limiting until about 135mA!