Regarding the 8060A and how the divider works:
In the 2V range, the input divider is in divide by 10 mode. This is 10M over 1.1111M. This divider is not used in DC, only AC. In DC the input is not divided for the 200mV and 2V ranges. However, in AC, the signal used is divided by 10 and has variable capacitors to adjust the response as flat as possible. So if the 2V range button is pushed, the input impedance is sitting at 11.1111M. So the 100M is switched in to bring it back to 10M. Only the 2V range is modified by the 100M resistor. I should note that the impedance is not exactly 10M in any range. The 2V, because of the 100M resistor comes the closest. In most ranges the input is slightly higher than 10M. In the 200mV range, the input impedance is 10.010M. In 2V, 10M (but the 100M resistor might be a 5%). The 20V range (divide by 100) has the biggest 10M variation; 10.101Mohms, ~1% high. The 200V and 1000V ranges use divide by 1000. The input impedance is 10.010M, the same as the 200mV range.
Whoops, I forgot the input protection resistor R2 adds 1kohm ±5% to all the above calculations. So the numbers should be: 200mV, 2V, 200V and 1000V ranges are 10.011M, 2V 10.001M (without counting the tolerance of the 100M resistor, R3), 20V is 10.102 Mohms.
These impedance variations are within the usual specified range of ±1% of most High Voltage probes. Ironically, many of the more modern Autoranging Fluke meters jump to 11.111M on their 3 or 4V ranges, so I guess the marketing group didn't care anymore. The instruction manual for the Fluke 80K-40 probe suggests a correction factor for those Multimeters. There are also high voltage probes that are for high impedance DMMs. These would work with the 8060A in its high impedance 200mV and 2V ranges (both function switches popped out) but if it is a divide by 1000 probe, you'd only be able to measure to 2000V.