Author Topic: Finally! Hantek HDM3065 6.5 Digit DMM  (Read 62908 times)
Offline infino
Re: Finally! Hantek HDM3065 6.5 Digit DMM
« Reply #125 on: May 03, 2025, 12:42:02 pm »
I've posted the link for the FW above, just check the earlier messages.
Thank you, I saw your files, but I thought there might be something more recent, and my multimeter didn’t want to update. The problem turned out to be that it requires a 2GB flash drive — it could see my 8GB one, but it wouldn’t update from it. And nowadays, it’s not so easy to find a 2GB flash drive )).
I updated the device, everything is working perfectly. Thank you for your help
Offline infino
« Reply #126 on: February 15, 2026, 11:44:50 pm »
After almost a year of using the device, it stopped saving settings and screenshots to an external USB flash drive. When a flash drive is connected, the device reports that a U-disk is connected, but it does not save any data. It also does not save measurement results or settings to the internal memory. Has anyone encountered a similar issue? Thank you.
The following users thanked this post: TUMEMBER
Offline infino
« Reply #127 on: March 18, 2026, 08:22:21 am »
I’ll answer my own question. For those who have this device, it’s a good idea to save the calibration settings to a USB drive for future use. To do this, insert a USB drive with a capacity of up to 2 GB, then press the Shift button followed by OK (you can also try holding the buttons). The settings will be saved to the USB drive as .ini files.
The manufacturer also provided software with firmware for recovery. To put the device into boot mode, with the device powered off, hold the DCV button and the “-” key, then turn it on — it will enter boot mode. After that, connect it to a computer via USB and run the manufacturer’s software. The script mfgtool2-yocto-mx-evk-nand.vbs will start, and the program will automatically reflash the NAND memory. Once the process is complete, the device will be restored.
After that, you should install the latest firmware and restore the calibration settings or recalibrate the device
The following users thanked this post: Microcheap
Offline infino
« Reply #128 on: March 21, 2026, 06:12:22 pm »
Hello everyone who’s reading this—though for now I’m replying to myself.
As I wrote above, the device support team sent me software that helped solve the problem, and they also provided a new firmware version: htdmm55_V2004(2510111618)_20251011.up. Unfortunately, with the new firmware, the program at the following link stopped working:
https://lygte-info.dk/project/TestControllerIntro%20UK.html
This is probably the only decent software that previously worked with the HDM3055. After reverting to the previous firmware, compatibility with the program was restored.
I asked the support team whether they plan to fix this issue, and received the following response from Hantek: “No compatible firmware available; we have no plans to fix this issue for now.”
It’s quite disappointing to see this level of support for the device. I’ll probably start looking for a replacement, although it’s hard to find something comparable at Hantek’s price point.
Offline Microcheap
« Reply #129 on: March 22, 2026, 05:05:42 am »
Hantek has no relation whatsoever with TestController, it's very likely that they don't even know what it is so, no surprise that they can't help.
I made the HDM3000 configuration for TestController. So maybe I can help.
How are you trying to connect? Serial or LAN? Can you connect to the multimeter using the Keysight Connection Expert?
If so, what's the reply for the command "*IDN?" ? It should be something like "Hantek, HDM3055, CNxxxxxxxxxxxxx"
« Last Edit: March 23, 2026, 08:46:03 pm by Microcheap »
Offline infino
« Reply #130 on: March 25, 2026, 02:08:33 am »
I understand that Hantek is not related to TestController. The issue is that their original software, HDM3000 Ver1.7, does not work correctly with the device — the panel always shows “Overload” (in all versions).
TestController used to work perfectly, but after the latest update it stopped functioning.
Here is the connection log when it does not work:
; jSerialComm version: 2.11.2
;; Start thread for: COM1
;; Start thread for: COM10
;; COM1: Set params: 9600
;; COM10: Set params: 9600
;; COM1: Tx: <*IDN?.> 2A 49 44 4E 3F 0A
;; COM10: Tx: <*IDN?.> 2A 49 44 4E 3F 0A
;; COM10: Rx: <Hantek,HDM3055,CN214xxxxxxxxxxxx,2.0.0.4>
;; USB Serial Converter (COM10) Did not find any match for: Hantek,HDM3055,CN214xxxxxxxxx,2.0.0.4
;; COM10: Close
;; Stopping thread for: COM10
And here is the connection log when everything works (with the previous firmware version of the device):
;; jSerialComm version: 2.11.2
;; COM10: Rx: <Hantek, HDM3055, CN214xxxxxxxx, 2.0.0.4>
;; COM10: Tx <*RST>
;; COM10: Tx <*CLS>
;; COM10: Tx <*OPC?>
;; COM10: Rx: <1>
;; Found Hantek HDM3055 on USB Serial Converter (COM10) sn: CN214xxxxxxxx
;; HDM3055: Tx SENSe:FUNCtion?
;; HDM3055: Rx <"VOLT">
ps I have a version of the device with a COM port and USB, but unfortunately no LAN. I cannot connect to TestController via USB — I’m connecting through the COM port, as shown in the terminal log.
« Last Edit: March 25, 2026, 02:15:38 am by infino »
Online KungFuJosh
  • TEAS is real.
« Reply #131 on: March 25, 2026, 02:18:53 am »
Do you have Scan serial ports checked in the Load devices tab? If so, disable it.
If something changed on the device, and TC stopped detecting it correctly, remove the device from TC and add it again fresh and see if that fixes the problem.
Thanks,
Josh
Best Continuity Tester Ever
Offline Kean
    • Kean Electronics
« Reply #132 on: March 25, 2026, 02:35:21 am »
It looks like the response to the *IDN? query no longer has the spaces after the commas.
Look closely at the differences in the two responses below.  That should be easy to fix in the TestController definition.
Offline infino
« Reply #133 on: March 25, 2026, 02:50:12 am »
I’ll be working tomorrow and will turn on the device to double-check the logs — maybe some spaces appeared when copying the text.
Offline Kean
    • Kean Electronics
« Reply #134 on: March 25, 2026, 02:56:26 am »
No, it looks like Hantek removed the spaces in the newer firmware.  They ideally should not have been there in the first place.
Offline Microcheap
« Reply #135 on: March 25, 2026, 06:12:31 am »
Their software has a bug and won't work correctly if the locate settings in windows is set to use comma as decimal separator (their power supply software has the same issue): https://www.eevblog.com/forum/testgear/hantek-hdm3055/msg5882732/#msg5882732
Regrading the connection with TC, Kean already pointed the issue. Apparently Hantek changed the response for the query *IDN? in this last firmware. The fix is very simple:
 - Look for the file "HantekHDM3000.txt" in the folder "../TestController/Devices/" and search for the line (30) "#idString Hantek, HDM3055," in this file. Now change it to match the new reply of your dmm: "#idString Hantek,HDM3055,". Save it and restart TestController.
It should connect now.
Offline infino
« Reply #136 on: March 25, 2026, 10:07:22 am »
I fixed it, but it didn’t help.
Starting
;; COM10: Rx: <Hantek,HDM3055,CN214xxxxxxxx,2.0.0.4> 48 61 6E 74 65 6B 2C 48 44 4D 33 30 35 35 2C 43 4E 32 31 34 30 30 33 30 30 30 34 36 32 2C 32 2E 30 2E 30 2E 34
;; USB Serial Converter (COM10) Did not find any match for: Hantek,HDM3055,CN214xxxxxxxx,2.0.0.4
;; COM10: Close
;; COM10: Close
;; COM1: Rx: timeout
;; COM1: Close
In Agilent-HP emulation mode.
Starting
;; COM10: Rx: <HEWLETT-PACKARD,34401A,0,9-4-1> 48 45 57 4C 45 54 54 2D 50 41 43 4B 41 52 44 2C 33 34 34 30 31 41 2C 30 2C 39 2D 34 2D 31
;; COM10: Tx <SerialInit>
;; COM10: Script start
;; COM10: Tx <SYST:REM>
;; COM10: Tx: <SYST:REM.> 53 59 53 54 3A 52 45 4D 0A
;; COM10: Delay: 20ms
;; COM10: Script end <>
;; Found Agilent 34401A on USB Serial Converter (COM10)
;; COM1: Close
;; HP34401A: Tx <FUNC?>
;; COM10: Tx: <FUNC?.> 46 55 4E 43 3F 0A
;; COM10: Rx: <"VOLT"> 22 56 4F 4C 54 22
;; HP34401A: Rx <"VOLT">
;; HP34401A: Tx <READ?>
« Last Edit: March 25, 2026, 10:52:30 am by infino »
Offline infino
« Reply #137 on: March 25, 2026, 11:03:35 am »
Everything is working now — it was my mistake. After editing, I didn’t save the file, and the editor was reading it from cache and showing it as if it had been modified.
Thank you for your help, and thanks to everyone who responded to the issue.
Offline likun123
  • Country: cn
« Reply #138 on: April 17, 2026, 04:22:50 am »
the latest official firmware for Hantek HDM3065.
htdmm65_V2006(2510231115)_2026041
Offline infino
« Reply #139 on: April 17, 2026, 12:46:32 pm »
I can’t find the update versions on the official website. Where do you check or find these versions?
Offline frankzoe
  • Country: cn
« Reply #140 on: April 19, 2026, 10:13:28 am »
This feature is very useful. I have successfully backed up the calibration files. In addition, a USB drive no larger than 2GB is not necessary; I use a 16GB USB 2.0 drive. Simply formatting the drive to the Fat32 file system allows it to work properly, whether for software upgrades or backing up calibration files.
Offline infino
« Reply #141 on: April 19, 2026, 12:59:03 pm »
Saving calibration files is definitely a very good practice. It’s a pity that I didn’t know about this option from the very beginning and only learned about it when Hantek sent me instructions on how to restore the device.
As for USB flash drive capacity: initially, my HDM3055 did not work with drives larger than 4 GB. More precisely, you could take a flash drive of any size, even 32 GB, create a 4 GB FAT32 partition on it, and then it would work. That’s exactly what I did.
After restoring the device and installing the latest software 2.0.0.4(2510111618), firmware 01.03.0601-22-02.05-02, I haven’t yet checked how it works with 8, 16, or 32 GB flash drives.
The main thing is that I managed to fully restore the device’s functionality — it now works perfectly with the testController software.