gslick:

--- Quote from: TK on May 11, 2017, 11:27:16 pm ---Do you have the schematics for 10300B Z80 preprocessor?  I am interested in knowing how to translate the control signals from Z80 to the 4-bit STAT bus that the Z80 inverse assembler needs.

--- End quote ---

The 10300B STAT signals in least significant order are:
/WR - CPU pin 22
/IORQ - CPU pin 20
/RFSH - CPU pin 28
/M1 - CPU pin 27

/WR is fed three times through a 74LS240 inverting buffer/driver for a STAT signal.
/IORQ is fed once through a 74LS240 inverting buffer/driver for clock signal, and twice for a STAT signal.
/RFSH is fed once through a 74LS241 non-inverting buffer/driver for a clock and STAT signal.
/M1 is fed once through a 74LS241 non-inverting buffer/driver for a STAT signal.

Maybe the multiple passes for /WR and /IORQ are to delay those signals by a few ns.

That is from the Model 64683A Interface Module Z80 Service Information manual part number 64683-90903. I only have a photocopy of the manual. I'll see if I can get it scanned when I have time.

TK:

--- Quote from: gslick on May 12, 2017, 02:34:39 am ---
--- Quote from: TK on May 11, 2017, 11:27:16 pm ---Do you have the schematics for 10300B Z80 preprocessor?  I am interested in knowing how to translate the control signals from Z80 to the 4-bit STAT bus that the Z80 inverse assembler needs.

--- End quote ---

The 10300B STAT signals in least significant order are:
/WR - CPU pin 22
/IORQ - CPU pin 20
/RFSH - CPU pin 28
/M1 - CPU pin 27

/WR is fed three times through a 74LS240 inverting buffer/driver for a STAT signal.
/IORQ is fed once through a 74LS240 inverting buffer/driver for clock single, and twice for a STAT signal.
/RFSH is once through a 74LS241 non-inverting buffer/driver for a clock and STAT signal.
/M1 is fed once through a 74LS241 non-inverting buffer/driver for a STAT signal.

Maybe the multiple passes for /WR and /IORQ are to delay those signals by a few ns.

That is from the Model 64683A Interface Module Z80 Service Information manual part number 64683-90903. I only have a photocopy of the manual. I'll see if I can get it scanned when I have time.

--- End quote ---
Thank you gslick.  This is the information I needed!!!

TK:

--- Quote from: gslick on May 12, 2017, 02:34:39 am ---
--- Quote from: TK on May 11, 2017, 11:27:16 pm ---Do you have the schematics for 10300B Z80 preprocessor?  I am interested in knowing how to translate the control signals from Z80 to the 4-bit STAT bus that the Z80 inverse assembler needs.

--- End quote ---

The 10300B STAT signals in least significant order are:
/WR - CPU pin 22
/IORQ - CPU pin 20
/RFSH - CPU pin 28
/M1 - CPU pin 27

/WR is fed three times through a 74LS240 inverting buffer/driver for a STAT signal.
/IORQ is fed once through a 74LS240 inverting buffer/driver for clock single, and twice for a STAT signal.
/RFSH is once through a 74LS241 non-inverting buffer/driver for a clock and STAT signal.
/M1 is fed once through a 74LS241 non-inverting buffer/driver for a STAT signal.

Maybe the multiple passes for /WR and /IORQ are to delay those signals by a few ns.

That is from the Model 64683A Interface Module Z80 Service Information manual part number 64683-90903. I only have a photocopy of the manual. I'll see if I can get it scanned when I have time.

--- End quote ---
I was using: /M1 /REFR /IORQ //WR (/WR inverted) and feeding the clock signal directly from the Z80, but noticed that Z80 inverse assembler was asking for 2 clock signals and did not know how to generate them.  Will try the new interface.

gslick:

--- Quote from: TK on May 12, 2017, 12:28:10 pm ---I was using: /M1 /REFR /IORQ //WR (/WR inverted) and feeding the clock signal directly from the Z80, but noticed that Z80 inverse assembler was asking for 2 clock signals and did not know how to generate them.  Will try the new interface.

--- End quote ---

The 64683A / 10300B clock signals are:
/MREQ - CPU pin 19 - fed once through a 74LS240 inverting buffer/driver for clock J
/IORQ - CPU pin 20 - fed once through a 74LS240 inverting buffer/driver for clock K
/RFSH - CPU pin 28 - fed once through a 74LS241 non-inverting buffer/driver for clock L

The default 10300B state configuration is to clock on (falling edge of J OR falling edge of K).

To filter out refresh cycles the state configuration could be changed to clock on ((falling edge of J OR falling edge of K) AND L high)

TK:

--- Quote from: gslick on May 12, 2017, 09:36:17 pm ---
--- Quote from: TK on May 12, 2017, 12:28:10 pm ---I was using: /M1 /REFR /IORQ //WR (/WR inverted) and feeding the clock signal directly from the Z80, but noticed that Z80 inverse assembler was asking for 2 clock signals and did not know how to generate them.  Will try the new interface.

--- End quote ---

The 64683A / 10300B clock signals are:
/MREQ - CPU pin 19 - fed once through a 74LS240 inverting buffer/driver for clock J
/IORQ - CPU pin 20 - fed once through a 74LS240 inverting buffer/driver for clock K
/RFSH - CPU pin 28 - fed once through a 74LS241 non-inverting buffer/driver for clock L

The default 10300B state configuration is to clock on (falling edge of J OR falling edge of K).

To filter out refresh cycles the state configuration could be changed to clock on ((falling edge of J OR falling edge of K) AND L high)

--- End quote ---
Impressive... Thanks!