Performa01:
--- Quote from: Rich@RohdeScopesUSA on October 20, 2018, 03:07:22 am ---We typically recommend the same method Daniel suggested (smallest volt/div setting, AC RMS or std deviation or a vertical histogram) with one addition (which I think Daniel would agree with) - you should really figure it as a percent of full scale as some oscilloscopes have 10 vertical divisions and some only have 8.
--- End quote ---
This is why I usually express the noise in LSB instead of µV in my graphs.
Of course, for this you need to know how many LSB/div. a certain DSO has - and keep in mind that this number will be far less for the zoomed (not full resolution) gain settings that some scopes have.
maxwell3e10:
Here is a compilation of noise spectra for several more oscilloscopes:
R&S RTB2004 thanks to Hydron
Instek GDS2204E thanks to nctnico
R&S RTM3004 thanks to nctnico
Rigol DS6104
I grouped them by the sampling rate, 1 GHz and 5 GHz. It looks like Rohde & Schwarz scopes have the lowest noise among the scopes tested so far.
Andreas:
Hello,
how many samples are you actually using for calculation for the 1 GS/s sample rate?
the minimum calculated frequency seems to be ~500 kHz with 500 kHz step resolution.
So it seems that you are using 2048 points for spectrum calculation.
(that would be a number which can easily uploaded as CSV here against the 1MS limitation from above).
with best regards
Andreas
Performa01:
Yes, for these graphs the data can easily be uploaded in this forum and that has been done in that other thread, linked in the opening post. Just have a look there.
The problem only arises when we shall provide data suitable for analysis down to 10Hz.
nctnico:
It would be interesting to see what the new DSO/MSO7000 from Rigol looks like when it comes to noise.