could you try this:
Select Voltage measurement, autorange and keep probes open.
Then slide your finger from left down corner to up.
Does it affect measurement result?
My 867 shows voltage jumps up to 0.03 V.
The same thing happens on R measurement, for 1 MOhm it may show even 0.3 MOhm at the moment of slide.
Is this ok?
I wrote to brymen support some time ago and they told me the following:
normal phenomenon. It is because the test current for 6MOhm range is less than around 0.125uA. For such kind of low current, the measurement may be easily affected by noise. If it gets 1Mohm reading while you do not touch plastic glass, then it should be no problem.
Also, when I slide finger from the left down corner to up, I see liquid crystal flashes on the display.
It seems, like high static electricity is induced on the other side of plastic glass.
May be there is need to use antistatic spray on the back side of the plastic glass?