Since we already have a few pics above, I'll just go over a few things for now. I measured a 10V reference I have, which my HP 3468B read as 10.004V. The B35 read it as 10.03V
The accuracy spec for DC voltage is +/- (0.5% + 2 counts). So for this reading it would be +/- .05V + 2 counts, so on that scale it would then be +/- .07 V so it's within the spec.
On the faceplate it also states that -
20A Max
Each 15 min Max 10 sec. Meaning, I think, that it can measure 20A Max for 10 sec but only once every 15 min.
Capacitance and resistance measurements looked good.
Is it cheaply made? Well, I've seen worse I guess but it's OK for the price so far. It came with a nice carrying bag and leads/attachments such as a K thermocouple and an attachment to assist in measuring caps/resistors and transistors.
I only played with the Bluetooth for a few minutes today but it was very easy to connect to my iPhone. I just watched the DC measurements draw a straight line, but will test more tomorrow. It does have a voice annunciation that states the reading each time if you wish it to, but my cat hated to listen to it. I need to get into the interface more tomorrow.