Author Topic: Pendulum CNT-90 Service Manual?  (Read 2490 times)
Offline perdrixTopic starter
« on: July 11, 2020, 05:38:45 pm »
Does anyone have this or know of its whereabouts?  It may also lurk under the name of the Fluke PM6690 or the Tektronix FCA3000 which were badge engineered versions.
Offline coromonadalix
Re: Pendulum CNT-90 Service Manual?
« Reply #1 on: July 11, 2020, 06:15:14 pm »
you have some infos in the fluke section,  not sure for any SV manuals
http://www.ko4bb.com/getsimple/index.php?id=manuals&dir=Fluke
Offline YetAnotherTechie
  • Country: pt
« Reply #2 on: July 13, 2020, 02:07:05 pm »
this units also exists under the fluke and the philips brand-names, one of them has a complete service manual, i found 2 revisions, and there are two hardware versions(each manual has both so no prob), you can find some info using the wayback machine too. i can't help you now, i can't find my files...
« Reply #3 on: July 13, 2020, 02:27:30 pm »
https://manualzz.com/doc/6289808/pm6690---fluke
a 9meg pdf can be downloaded
Hope it helps
Uploaded to ko4bb,  may take a while to appear ??
« Last Edit: July 13, 2020, 02:33:07 pm by coromonadalix »
The following users thanked this post: Teti, notfaded1
« Reply #4 on: August 07, 2020, 08:21:58 am »
Now at ko4bb
section
Philips/PM6690_Frequency_Counter
Offline Teti
  • Country: cz
    • www.martenelectric.cz
« Reply #5 on: September 09, 2020, 07:07:15 pm »
Hope it helps
Thank you  :)
Vintage audio gear connoisseur, computer enthusiast, time-nut, music lover, vintage games gamer, nerd, tinkerer and shady electronic projects maker.
http://www.martenelectric.cz/
Offline Teti
  • Country: cz
« Reply #6 on: September 09, 2020, 07:21:25 pm »
Do you by any chance know what is a calibration password?
Offline DonRon
« Reply #7 on: September 09, 2020, 08:21:48 pm »
Hi Teti,
found the service-manual for Fluke PM6690:
Have a look at page 7-3 The Calibration Submenu ...
"The Calibration Submenu
When used for the first time, this submenu can be activated
from the USER OPT menu after first entering the fixed pass
-word 62951413. All calibrations that can be performed manu
-ally are controlled from here. "
So maybe 62951413 might work ..
Cheers,
Ronald
Offline noreply
« Reply #8 on: September 10, 2020, 10:52:11 pm »
Played around with an original CNT-90 some time back - can definitely confirm this code is  the default PW :-+
Offline Teti
  • Country: cz
« Reply #9 on: September 16, 2020, 02:41:49 pm »
Thanks guys, cal password worked a treat!