joeqsmith:
--- Quote from: SJL-Instruments on February 03, 2024, 02:22:33 pm ---Luckily, this cannot happen, since the returned values are expressed as an integer fraction over 255 (odd). :)
--- End quote ---
:palm: :-DD
So you want it coded as written where a value of 0.5 is included in both bins? So 3 points minimum?
SJL-Instruments:
--- Quote from: joeqsmith on February 03, 2024, 02:28:38 pm ---So you want it coded as written where a value of 0.5 is included in both bins?
--- End quote ---
In our software, we check for 2 points satisfying 0.5 < x < 0.9 and 2 points satisfying 0.1 < x < 0.5.
Whether you include the endpoints has no effect on the behavior, since none of the values 0.1, 0.5, 0.9 can occur exactly given the serial encoding scheme.
joeqsmith:
1.1 states the absolute input voltage is +/-1.2V or 2.4V full scale. 1.5 states vertical resolution is 12-bits or 560uV. With this in mind, showing a 10MHz sine applied, 5ps resolution. Please explain the steps.
SJL-Instruments:
--- Quote from: joeqsmith on February 03, 2024, 05:21:38 pm ---1.1 states the absolute input voltage is +/-1.2V or 2.4V full scale. 1.5 states vertical resolution is 12-bits or 560uV. With this in mind, showing a 10MHz sine applied, 5ps resolution. Please explain the steps.
--- End quote ---
We cannot reproduce these steps with the official software, looking at a similar timebase/voltage range and with a 10 MHz sine wave. (See attached screenshot.)
Can you see the steps using our software?
joeqsmith:
Yes, I can show it with your software but the difference is the source. We are seeing effects from my old Marconi 2024. Time to ditch it and go with the old HP8640B cavity.