SJL-Instruments:

--- Quote from: joeqsmith on February 08, 2024, 04:50:01 am ---When looking at the start of collection, what determines the minimum value before it reports an underflow? The manual states:

--- Quote ---Error Response: WARNING: Delay underflow -inf  This error occurs when the requested value is outside the range of the delay generator and cannot be achieved.  If you are certain the requested delay falls within acceptable bounds, contact us for
a diagnosis.
--- End quote ---


Above it states:

--- Quote ---Undefined behavior may occur if the requested delay is lower than 11 ns (D110000) or higher than the holdoff minus 25 ns. It is the user’s responsibility to stay within these bounds.

--- End quote ---

The term bounds must be referring to different boundary conditions as it does not appear to be constant for a fixed holdoff time.  Can you provide the equations?   7.7ns seems to work in all of my test cases so far. 

The document does mention the delay generator in 2.3 but doesn't mention how it ties into the underflow warning. 

--- End quote ---
The minimum delay is determined by the combined gate delays of all chips between the CH1 input and the sampling circuit. This is unit-dependent, and there isn't a useful way to calculate it. The easiest method is to simply try smaller delays until you receive an underflow warning (as you've done).

The maximum delay is determined by the 25 ns reset time of the sampling circuit. If less than 25 ns is allocated between the sampling time and the end of the trigger holdoff, the sampling circuit will be in an indeterminate state at the next trigger and yield incorrect results.



--- Quote from: joeqsmith on February 08, 2024, 04:50:01 am ---I suggest as you make changes to the hardware, you consider sending a set of replacement parts for this unit anyone working on a review so they remain current.   For a while, I was thinking to have you ship a production unit that you bless as a gold standard to have as a reference.  I no longer feel that will be useful.   After all of my testing and our conversation, I'm now fairly confident that the electronics on this particular unit represent production. 

--- End quote ---
We are also confident, given the last several posts, that your unit is within production spec.
We can certainly send you an updated case, if you'd like. (We did recently send a revised version to Shahriar.)

joeqsmith:

--- Quote ---The minimum delay is determined by the combined gate delays of all chips between the CH1 input and the sampling circuit. This is unit-dependent, and there isn't a useful way to calculate it. The easiest method is to simply try smaller delays until you receive an underflow warning (as you've done).

--- End quote ---

Could you explain the actual hardware/firmware mechanism in further details which sets this warning?  Block diagram of that area of the hardware design? 

For keeping the unit up to date, I would suggest anything that effects the performance of the scope, automatically send it.   I'm not concerned with the cosmetics.  If you do decide to start including probes, I suggest sending these out to reviewers as well.   


Besides decreasing the sampling resolution, is there a way to speed up the sweeps while reducing performance?  What limits the sweep rates? 

I have not yet measured it but it seems pulling all four channels vs one makes little difference.  Same with reducing the number of CDFs.  Would adding a mode where you tell the firmware/hardware to sweep a range and just have it stream the data back rather than requesting each delay, speed up the process?   I'm not thinking small changes in sweep times but 10X, 100X faster.   

If do decide to go after some sort of auto setup mode, it would be nice if it not only set the vertical and horizontal but also the trigger level.  I have not attempted to automate this myself.

***
I checked the Signal Path channel yesterday to see if he had his review up yet.   I keep postponing mine as you are making so many improvements in such a short time,  by the time I release a review, it would be obsolete.   I would have a different view if you were not using the feedback I am providing and working at such a rapid pace. 

SJL-Instruments:

--- Quote from: joeqsmith on February 08, 2024, 02:33:23 pm ---
--- Quote ---The minimum delay is determined by the combined gate delays of all chips between the CH1 input and the sampling circuit. This is unit-dependent, and there isn't a useful way to calculate it. The easiest method is to simply try smaller delays until you receive an underflow warning (as you've done).

--- End quote ---

Could you explain the actual hardware/firmware mechanism in further details which sets this warning?  Block diagram of that area of the hardware design?

--- End quote ---
 
Essentially, the entire delay generator is looped back on itself to form a ring oscillator, and we measure the period to verify the delay. If the calibration is still valid, this should be correct to within 1 ps. If it differs from the target by more than 2 ps, the warning is issued.
If the requested delay is too small, the delay generator will clamp to its minimum delay. This will be caught by the above check, but where this occurs is subject to process variation.


--- Quote from: joeqsmith on February 08, 2024, 02:33:23 pm ---For keeping the unit up to date, I would suggest anything that effects the performance of the scope, automatically send it.   I'm not concerned with the cosmetics.  If you do decide to start including probes, I suggest sending these out to reviewers as well.   

--- End quote ---
We'll ship you an updated case by EOD tomorrow. Probes are quite far from production, but we'll keep you updated.


--- Quote from: joeqsmith on February 08, 2024, 02:33:23 pm ---Besides decreasing the sampling resolution, is there a way to speed up the sweeps while reducing performance?  What limits the sweep rates? 

I have not yet measured it but it seems pulling all four channels vs one makes little difference.  Same with reducing the number of CDFs.  Would adding a mode where you tell the firmware/hardware to sweep a range and just have it stream the data back rather than requesting each delay, speed up the process?   I'm not thinking small changes in sweep times but 10X, 100X faster.   

If do decide to go after some sort of auto setup mode, it would be nice if it not only set the vertical and horizontal but also the trigger level.  I have not attempted to automate this myself.

--- End quote ---
First, there is the theoretical minimum time needed to take the CDF samples themselves. At 40 ns holdoff, with 20 samples/CDF and 100 triggers/sample (which would give degraded but still "acceptable" performance), this is 80 us per point, corresponding to 12 kCDF/s.

There is a fixed minimum overhead per sample due to FPGA/DAC communication, analog settling time, and MCU processing time. With some optimization, this can be brought down to probably ~10 us per sample, so ~200 us per point.

Then there is the time required to set a delay. Right now this takes ~8 ms, and gives a precision of ~0.1 ps @ 20 ns delay.
We can cut this down to ~50 us by skipping the verification step. This will substantially degrade the ENOB.

Combining the above, this limits the speed to ~3 kCDF/s.

Each CDF requires 60 bytes. Given the 921600 baud rate, this limits the speed to 1920 kCDF/s. Many of the CDF entries are strictly-speaking, redundant, and with pruning we can probably achieve 3 kCDF/s.
The current speed with default settings is ~50 CDF/s. This will take some time to develop, but a factor of ~100x is possible, with degraded noise and jitter performance.


--- Quote from: joeqsmith on February 08, 2024, 02:33:23 pm ---I checked the Signal Path channel yesterday to see if he had his review up yet.   I keep postponing mine as you are making so many improvements in such a short time,  by the time I release a review, it would be obsolete.   I would have a different view if you were not using the feedback I am providing and working at such a rapid pace. 

--- End quote ---
We appreciate the amount of feedback you've provided, and the time you've put into testing the scope. Keep in mind that we won't stop improving the product for the foreseeable future - be careful not to postpone the review indefinitely.  ;D

joeqsmith:

--- Quote ---Essentially, the entire delay generator is looped back on itself to form a ring oscillator, and we measure the period to verify the delay. If the calibration is still valid, this should be correct to within 1 ps. If it differs from the target by more than 2 ps, the warning is issued.  If the requested delay is too small, the delay generator will clamp to its minimum delay. This will be caught by the above check, but where this occurs is subject to process variation.

--- End quote ---

Thanks.   Makes sense why you need to constantly align it and why it changes a few counts.     

I'll check the isolation again once the new parts arrive and post any improvements.   Obviously I had the PCB out of the case to perform the rework.  The thermal pads appear to still be fine if you did not want to include them. 


--- Quote ---We appreciate the amount of feedback you've provided, and the time you've put into testing the scope. Keep in mind that we won't stop improving the product for the foreseeable future - be careful not to postpone the review indefinitely.  ;D

--- End quote ---
:-DD   Are you getting any feedback from Signal Path? 

SJL-Instruments:

--- Quote from: joeqsmith on February 09, 2024, 12:09:23 am ---I'll check the isolation again once the new parts arrive and post any improvements.   Obviously I had the PCB out of the case to perform the rework.  The thermal pads appear to still be fine if you did not want to include them.

--- End quote ---
We have included new pads with the revised case. They are ~1 mm thick, and will compress significantly upon first assembly to obtain best thermal contact. Will be shipped out tomorrow.


--- Quote from: joeqsmith on February 09, 2024, 12:09:23 am --- :-DD   Are you getting any feedback from Signal Path? 

--- End quote ---
We can't comment on that, sorry. (Not sure whether Shahriar wants his schedule publicized.)